I have a method that no matter what I do I seem to get persistent carryover that doesn't reduce after multiple diluent injections. Has anyone come across materials binding to the rheodyne and then being washed off at a later time? Any suggestions?
Find maintenance, best practice, and troubleshooting articles.
I have a method that no matter what I do I seem to get persistent carryover that doesn't reduce after multiple diluent injections. Has anyone come across materials binding to the rheodyne and then being washed off at a later time? Any suggestions?
Hi,
We are running on 1200 and 1260, quaternary pump, DAD detectors.
The compound of interest in this case is Trifluoperazine HCl, running phosphate buffer with triethyamine and ACN gradient.
The column is an Agilent Poroshell 120 EC-C18 2.7µm 50 x 4.6mm
The method has been validated for product license requirements.
The issue was seen during the development and validation process intermittently however we were never able to get to the bottom of the issue.
Sometimes we don’t see any ‘carryover’, sometimes we see a small amount, this time I have seen ~1%
I’m at a bit of a loss to explain it so welcome almost any suggestions.
Kat
HI,
Just this compound, though there are multiple compounds in the solutions.
And no, both diluent injections and blank gradient runs don’t show significant reduction of the peak in question.
If we very thoroughly clean the system, this seems to improve the amount of the peak present, but it does seem to increase again fairly rapidly. We can sometimes squeeze out a run before the peaks gets problematically big.
The development analyst tried multiple syringe washes including IPA to clean up between injections with no luck
Kat
Inj vol 5µm
Area in Std – 16
Area carried into diluent – 0.15
Concentration of std – 0.1mg/ml
Thanks both,
Ill check this out tomorrow when I’m back in the Lab.
Can you give us some details about that modules you have? What is the method? What are the compounds of interest? What column are you using?
A little bit of details will help us figure out what you are trying to do and give better ideas
Does the carry over affect all compounds or just this compound?
Do you see the carry over when you run blank injections?
If you do an injection and then run multiple blanks after, do you see it in all blanks? and does it seem to go down?