Why does RID signal shut down for no apparent reason?

Hi,

I've been experiencing some issues with a RID, used as the only detector. At any moment, and for no apparent reason, the RID stops (yellow label), displaying in the tooltip a "balancing error". When clicking then on "Balancing", it doesn't help, but opening the purge valve for a few minutes fixes the problem (green label back). This happens more and more, at the start or during a run. 

I use HPLC grade water as the eluent, and the temperature of the RID is 40 °C. 

Has someone come accross a similar problem?

Thanks!

Parents
  • Thank you for your answer. Indeed, more precisely the message is "unbalanced diodes".

    This happens mainly in two different situations : when starting the RID after some days off, and during the run of a sample.

    The samples are not over-concentrated. Moreover, it seems to happen more and more frequently.

    Would the warning message be just normal or would something cause an increased sensitivity ? Could something be done to avoid this ?

    Thank you again.

Reply
  • Thank you for your answer. Indeed, more precisely the message is "unbalanced diodes".

    This happens mainly in two different situations : when starting the RID after some days off, and during the run of a sample.

    The samples are not over-concentrated. Moreover, it seems to happen more and more frequently.

    Would the warning message be just normal or would something cause an increased sensitivity ? Could something be done to avoid this ?

    Thank you again.