G1362A RID Diode Balance. -1 or 1

Hello, while trying to balance the diodes in our RID we discovered that no matter how small of an adjustment is made the signal will only vary from -1 to 1.  The detector has been purged for hours with water and mobile phases and appears to be free of any air bubbles.  My thoughts are that one of the light receiving diodes has become faulty thus making balance impossible.  Any help would be greatly appreciated.  

Parents Reply
  • Azrael, honestly, it could be any kind of contamination or damage in/on the cell. Of course, if someone lets the system sit in NaOH or something similar, the cell will suffer.

    As mentioned earlier: It's vital to thoroughly flush both sides of the cell with water and then with IPA (obviously not with water, if normal phase was done before) when the RID is not in use or even gets stored for a longer period of time.