1290 not connecting

Our 1290 was working fine until we moved it a couple of feet to another bench. Now, the modules don’t want to connect, they show yellow lights upon startup.

Going into LabAdvisor, it says “Could not connect to 1290. Ensure that the system is switched on and not connected by another controller”

  • We only disconnected a couple of the CAN cables, and the ethernet to the computer. These are reconnected.
  • The CAN cables are connected to CAN sockets, and I verified they are seated.
  • Ethernet is connected from the DAD to the computer, and is plugged into a LAN connection on the detector, I verified it wasn’t a CAN connector

Please advise what else I should look for.

Thanks in advance!

Parents
  • So, if all DIP switches are down and you have an Infinity II DAD, the mode it is in, is called "using default", so the IP address should be 192.168.254.11. However, if, for whatever reason, your PC is not in the same IP range (e.g. 192.168.254.1), you will never be able to successfully ping the DAD. Therefore I suggested to also ping your PC, not only the DAD. Of course, the subnet mask on the PC should also be identical to the setting on the DAD (255.255.255.0).

     

    If it's still not working (also not with another LAN or crossover cable), do a forced cold start on the DAD - it's not difficult at all:

    - Switch off the DAD

    - Set DIP switches 1 and 6 to the upper position (carefully use a small flat head screwdriver or a tooth pick)

    - Switch on the DAD and wait 1 min

    - Switch off the DAD

    - Put all DIP switches to their original position (e.g. all down=using default)

    - Switch on the DAD

     

    See if you can now ping the DAD.

     

    It might be useful to borrow a G4208A Instant Pilot from another LC and connect it to the DAD to find more details about the current IP status.

  • Resetting the DAD worked! After that, the PING worked, followed by Lab Advisor, and lastly, the components on OPenLab are showing a reasonable status (idle).

     

    Thanks much!