i've changed frit and golden seal cap of my HPLC Agilent 1260 as ususal. This time, after the replacement of the valve, there's a leak, I' ve tighten the valve but there is a drop: what can I do?
thanks!
Find maintenance, best practice, and troubleshooting articles.
hello,
i've changed frit and golden seal cap of my HPLC Agilent 1260 as ususal. This time, after the replacement of the valve, there's a leak, I' ve tighten the valve but there is a drop: what can I do?
thanks!
Hi,
could you please provide the nut part number and how to replace the purg valu nut
Hello Kancheri
this is the purge valve for Agilent 1260 G1312-60071 | Agilent
You can check in the manual at page 154 on HOW to replace purge valve with this document
https://www.agilent.com/cs/library/usermanuals/Public/G1312-90015_BinPump-B_USR_EN.pdf