Purge  valve leaks_AGilent 1260

hello,

i've changed frit and golden seal cap of my HPLC Agilent 1260 as ususal. This time, after the replacement of the valve, there's a leak, I' ve tighten the valve but there is a drop: what can I do?

thanks!

Parents
  • Hello, this is the purge valve you have installed on the head of the pump? If yes, I usually never change the gold seal, but I only take it out for the ptfe frit change, then I'm reistalling the same gold seal.

    If the leak is on

    A : replace the nut and the tubing

    B : probably the gold seal is not perfectly adapted, change it again.

    B "MacGyver version" : use teflon on the thread...at least you won't see any drop  

     

    Alway, never overtight it, it get stuff worse.

     

    Hope you solve the problem.

Reply
  • Hello, this is the purge valve you have installed on the head of the pump? If yes, I usually never change the gold seal, but I only take it out for the ptfe frit change, then I'm reistalling the same gold seal.

    If the leak is on

    A : replace the nut and the tubing

    B : probably the gold seal is not perfectly adapted, change it again.

    B "MacGyver version" : use teflon on the thread...at least you won't see any drop  

     

    Alway, never overtight it, it get stuff worse.

     

    Hope you solve the problem.