hello,
i've changed frit and golden seal cap of my HPLC Agilent 1260 as ususal. This time, after the replacement of the valve, there's a leak, I' ve tighten the valve but there is a drop: what can I do?
thanks!
Find maintenance, best practice, and troubleshooting articles.
hello,
i've changed frit and golden seal cap of my HPLC Agilent 1260 as ususal. This time, after the replacement of the valve, there's a leak, I' ve tighten the valve but there is a drop: what can I do?
thanks!
Hello, this is the purge valve you have installed on the head of the pump? If yes, I usually never change the gold seal, but I only take it out for the ptfe frit change, then I'm reistalling the same gold seal.
If the leak is on
A : replace the nut and the tubing
B : probably the gold seal is not perfectly adapted, change it again.
B "MacGyver version" : use teflon on the thread...at least you won't see any drop
Alway, never overtight it, it get stuff worse.
Hope you solve the problem.
I have the same problem and did not know how can I resolve it
Hi
What is this yellow color?
If you are changing frit and gold cap then before installing dip it in mobile phase for at least 5 minutes in beaker.
After this, you can replace frit n gold cap..
Hi
What is this yellow color?
If you are changing frit and gold cap then before installing dip it in mobile phase for at least 5 minutes in beaker.
After this, you can replace frit n gold cap..