Servo Error on 1260 Quaternary Pump

Thanks a lot for your help. I fixed the problems I had on the pump but new troubleshooting appeared when I started the device, please find attached the screenshot of different error messages displayed on the pump logbook via Agilent Lab Advisor software. I checked the head pump and the different cable with the mainboard but no result.

What do you suggest to fix this issue and how to solve this problem?

Best regards.

Parents
  • Hello,

     

    For the Servo Restart Failed error, it can be created from the pump drive not moving properly. It can also be cause by a blockage of the Inlet valve or Outlet valve. So you can try to loosen one of them and start the module and see if it gives the error again. Otherwise the issue can be pump drive needs to be changed or mainboard is defective.

     

    Motor Power Too High means that the power drive is drawing too much current. So the steps to verify what  the issue is are the same as above. Loosen valve, start module. If it still gives the error with both valves loose, it is either pump drive or mainboard again.

     

    Pump Head missing means that there was no pump head at the time and the module gave the error for that. Either that or you need to change your pistons. 

     

    Overall the issue seems to be either your pump head is blocked inside, wrongly put back together, pump drive is defective or your mainboard is defective. 

     

    I would suggest trying to do what you can and if it comes to pump drive or mainboard, calling in to get an engineer onsite to look at it for you.

Reply
  • Hello,

     

    For the Servo Restart Failed error, it can be created from the pump drive not moving properly. It can also be cause by a blockage of the Inlet valve or Outlet valve. So you can try to loosen one of them and start the module and see if it gives the error again. Otherwise the issue can be pump drive needs to be changed or mainboard is defective.

     

    Motor Power Too High means that the power drive is drawing too much current. So the steps to verify what  the issue is are the same as above. Loosen valve, start module. If it still gives the error with both valves loose, it is either pump drive or mainboard again.

     

    Pump Head missing means that there was no pump head at the time and the module gave the error for that. Either that or you need to change your pistons. 

     

    Overall the issue seems to be either your pump head is blocked inside, wrongly put back together, pump drive is defective or your mainboard is defective. 

     

    I would suggest trying to do what you can and if it comes to pump drive or mainboard, calling in to get an engineer onsite to look at it for you.