Loss of sensitivity after instrument vent and column change 7250 GC/Q-TOF

I vented and changed columns recently and have noticed at least a ten fold loss of sensitivity.  The retention times and peak shape seems to be similar to what I have had in the past. I ran the Air and Water Check and passed. I have no issues calibrating and the calibrant sensitivity is the same as it was before, so I expect the problem is with the sample getting to the MS? I'm using 2, 15m columns connected with a PSD.  Would there be any other diagnostics test that I should be using to trouble shoot? Thanks for any thoughts. 

  • How far is the column installed up into the injection port?  How far is the column installed out of the transferline tipseal into the MS?  When did you last do full injection port maintenance? In the autotune reports, is the filament current different before/after the columns were replaced?   Autotune adjusts parameters to get the results into programmed ranges and the algorithm compensates for changes.  Were both columns the same as the old ones?  Are they cut properly and have new ferrules at the Purged Ultimate Union?  Did you change the column configuration for the new ones?  Are the columns calibrated in MassHunter Acquisition?    Are the vacuum readings different than before?

    Could you share the contents of the acqmeth.txt file that is in your run method's Windows subdirectory along with a tune report?

  • Maintenance on the injection port was done a month ago and only around 50 injections since then. I went head and replaced the liner and double checked the column.  I noticed that the column did slip when being installed and was into the injection port about 4x as far as it should have been.  The filament current was the same before and after the column install.  The column is the same and new ferrules were used.  I did not calibrate the columns in MassHunter Acquisition.  The column is from Agilent and selected them from the "Select from Catalog" option in the method.  The vacuum readings are the same as before I vented.  I'm running a standard now and hope that changing the length of the column going into the injection port solves the problem.  Thanks

                          INSTRUMENT CONTROL PARAMETERS:    GC QTOF
                          -----------------------------------------
    
       D:\MassHunter\GCMS\1\methods\Pesticide_Split_50_Training101323.M
          Sun Nov 10 21:50:42 2024
    
    Control Information
    ------- -----------
    
    Sample Inlet             : GC
    Injection Source         : GC ALS
    Injection Location:  Front
    Mass Spectrometer        : Enabled
    
    
     No Sample Prep method has been assigned to this method.
    
    
    GC
    GC Summary
    Run Time                                     15.875 min
    Post Run Time                                2 min
    
    Oven
    Temperature
    Setpoint                                     On
    (Initial)                                    90 °C
    Hold Time                                    1 min
    Post Run                                     280 °C
    Program
    #1 Rate                                      10 °C/min
    #1 Value                                     225 °C
    #1 Hold Time                                 0 min
    #2 Rate                                      40 °C/min
    #2 Value                                     280 °C
    #2 Hold Time                                 0 min
    
    
    Equilibration Time                           1 min
    Max Temperature                              320 °C
    Maximum Temperature Override                 Disabled
    Slow Fan                                     Disabled
    
    ALS
    Front Injector
    Syringe Size                                 10 μL
    Injection Volume                             1 μL
    Solvent A Washes (PreInj)                    3
    Solvent A Washes (PostInj)                   3
    Solvent A Volume                             8 μL
    Solvent B Washes (PreInj)                    3
    Solvent B Washes (PostInj)                   3
    Solvent B Volume                             8 μL
    Sample Washes                                1
    Sample Wash Volume                           8 μL
    Sample Pumps                                 6
    Dwell Time (PreInj)                          0 min
    Dwell Time (PostInj)                         0 min
    Solvent Wash Draw Speed                      300 μL/min
    Solvent Wash Dispense Speed                  3000 μL/min
    Sample Wash Draw Speed                       300 μL/min
    Sample Wash Dispense Speed                   3000 μL/min
    Injection Draw Speed                         300 μL/min
    Injection Dispense Speed                     6000 μL/min
    Viscosity Delay                              2 sec
    Sample Depth                                 Disabled
    Injection Type                               Standard
    L1 Airgap                                    0.2 μL
    Solvent Wash Mode                            A,B
    
    Sample Overlap
    Mode                                         Sample overlap is not enabled
    
    ALS Errors                                   Pause for user interaction
    
    Front SS Inlet He
    Mode                                         Split
    Heater                                       On    250 °C
    Pressure                                     On    11.157 psi
    Total Flow                                   On    56 mL/min
    Septum Purge Flow                            On    5 mL/min
    Pre-Run Flow Test                            Off
    Gas Saver                                    On    15 mL/min after 2 min
    Split Ratio                                  50 :1
    Split Flow                                   50 mL/min
    Liner                                        Agilent 5190-2295: 870 μL (Split, taper, wool, low pressure drop)
    
    Column
    Column #1
    Flow
    Setpoint                                     On
    (Initial)                                    1 mL/min
    Post Run                                     -4.273 mL/min
    
    Column Information                           Agilent 19091S-431UI: us2702611a
    Description                                  HP-5ms Ultra Inert
    Temperature Range                            -60 °C—325 °C (350 °C)
    Dimensions                                   15 m x 250 μm x 0.25 μm (Uncalibrated)
    Heater                                       Oven
    Column lock                                  Unlocked
    In                                           Front SS Inlet He
    Out                                          PSD 1 
    (Initial)                                    90 °C
    Pressure                                     11.157 psi
    Flow                                         1 mL/min
    Average Velocity                             26.949 cm/sec
    Holdup Time                                  0.92769 min
    Control Mode                                 Constant Flow
    
    Column #2
    Flow
    Setpoint                                     On
    (Initial)                                    1.2 mL/min
    Post Run                                     4.712 mL/min
    
    Column Information                           Agilent 19091S-431UI: US2407041B-2
    Description                                  HP-5ms Ultra Inert
    Temperature Range                            -60 °C—325 °C (350 °C)
    Dimensions                                   15 m x 250 μm x 0.25 μm (Uncalibrated)
    Heater                                       Oven
    Column lock                                  Unlocked
    In                                           PSD 1 He
    Out                                          MSD 
    (Initial)                                    90 °C
    Pressure                                     4.3978 psi
    Flow                                         1.2 mL/min
    Average Velocity                             57.524 cm/sec
    Holdup Time                                  0.4346 min
    Control Mode                                 Constant Flow
    
    Column Outlet Pressure                       0 psi
    
    Detector Evaluation
    Perform Detector Evaluation Test             Off
    
    MSD Transfer Line
    Temperature
    Setpoint                                     On
    (Initial)                                    280 °C
    
    
    Collision Cell/Aux
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Quench Gas He                                On    4 mL/min
    Collision Gas N2                             On    1 mL/min
    Collision Cell Aux He                        (unused)
    
    PSD 1
    PSD Purge                                    On    5 mL/min
    PSD 1 He                                     Supplies Column 2
    
    Signals
    Signal #1: 
    Description                                  None
    
    Signal #2: 
    Description                                  None
    
    Signal #3: 
    Description                                  None
    
    Signal #4: 
    Description                                  None
    
    Signal #5: 
    Description                                  None
    
    Signal #6: 
    Description                                  None
    
    Signal #7: 
    Description                                  None
    
    Signal #8: 
    Description                                  None
    
    
    
    
    
    
                            END OF INSTRUMENT CONTROL PARAMETERS
                            ------------------------------------
    

  • It is rare that I have so few method recommendations!  Well done.  Please increase the Gas Saver flow to 20 ml/min or even 25 ml/min.  If that flow is too low there is a chance that air can diffuse into the inlet subsystem and damage the columns.

    That's a Mass Calibration Report and not an autotune.

    If the column is way too high in the liner especially in split mode, the sample can vaporize below the end and be mostly swept away. I'm a fan of about 8mm of column sticking up past the ferrule during column installation in a Split/Splitless injection port.

    Does your lab have two 7250s...US2003E302 - the one in this MassCal report and US2013E302 ?   

  • I attached the autotune file now, sorry about that.  I was using split mode so hopefully explains the loss of signal. We only have one instrument, and it is US2003E302. Thanks for the help

     

  • Thank you, that cleared up something that had bothered me since it was installed!

    Tune looks great. 