Hello there,
I am working with an Agilent GC 8890 system coupled to a 5977B GC/MSD. At first hour in the morning, when I do the autotune to the instrument, the nitrogen level is extremely high (<25 %). As a solution, I purge the system with a high flow of helium in the injection port (around 500 mL/min He), during 30 mins aproximately, and then I do the tune again, obtaining values of nitrogen around 5 % (acceptable). So, the system is suitbale to carry on an analysis. I programmed a secuence of two samples, with an analysis of one and a half hour of duration each (chromatographic separation of 30 mins). Then, after completing the secuence, I carry over again an autotune, obtaining a value of nitrogen over 35 %. Where do I have a leak? I don't have an universal trap. In previous analysis, I decided to take it off, because I thought it could be contaminated or had a leak. Instead, I only use a gas clean filter (which is brand new).
The chromatographic conditions are:
- Front inlet: 200 ºC; Split 15:1 (total flow 25 mL/min)
- Column: oven initial temperature: 30ºC
- Aux heater: 260 ºC
Thanks in advance.