NCI GCMS methane pre tune and PCI tune failing but NCI tune passing

We have a NCI GCMS 5977B. We are facing methane pre tune and PCI tune failing. It says, "Improper masses for mass axis calibration".
We installed EI source. Did auto tune and it passed.

After that, we cleaned CI source and installed back. Baked out for 12 hours, then tried methane pre tune, but its not completing.

Tried PCI tune, same error and its not completing.

Then tried NCI tune and its completing.

and we ran standard of MCCP, recovery is good. 

We tried with new CI filament, but same result. We tried with an EI filament also, but no change.

Please help to diagnose the issue.

Parents
  • Load a previous tune like one that worked some months ago.

     Or, go back to an older data file that has PCICH4.U underneath the subdirectories and load that one.

    The CI Wizard screen: "Enter Reagent Ion Masses for Flow Adjustment"  must be 17, 29, 41. 

    The PCI Tune Ions must be 41, 267, 599.

    If you load PCICH4.U and select the More... button

    The defaults are these:

    Good job trying EI, the EI filament in CI, and NCI to prove that the rest of the MS is working!

  • Hi Paulo, I have tried with previous good tune file, and also tried another file putting default values, but issue is same. Improper mass axis for calibration.

  • In EI, the Mass Gain and Offset are 0, -23.  In NCI, the Mass Gain and Offset are -4, -20 << almost the same as EI.

    In PCI when it fails, the Mass Gain and Offset are 2047, -481 -- so the mass assignment is very wrong for flow adjust with Methane as the reagent gas.  The PCI tune algorithm is trying to adjust the mass gain and offset for the largest peak in the window and drive it to the setpoint for the CI gas selected. Please make sure that your CI gas is configured to be Methane. If it's configured for Ammonia or Isobutane the algorithm will be trying to drive to those masses.

    Or, manually change the Mass Gain to 0 and the Offset to -20, and manually run the methane pre-tune to see if that finishes rather than running Autotune.

Reply
  • In EI, the Mass Gain and Offset are 0, -23.  In NCI, the Mass Gain and Offset are -4, -20 << almost the same as EI.

    In PCI when it fails, the Mass Gain and Offset are 2047, -481 -- so the mass assignment is very wrong for flow adjust with Methane as the reagent gas.  The PCI tune algorithm is trying to adjust the mass gain and offset for the largest peak in the window and drive it to the setpoint for the CI gas selected. Please make sure that your CI gas is configured to be Methane. If it's configured for Ammonia or Isobutane the algorithm will be trying to drive to those masses.

    Or, manually change the Mass Gain to 0 and the Offset to -20, and manually run the methane pre-tune to see if that finishes rather than running Autotune.