Good Morning,
We have a problem with one of our newer GC-MSs (5977B MSD with Turbo pump) that is about 2 years old. The turbo pump seems not to be able to kick in after the initial rough pump started.
The detailed problem is described below:
One day we suddenly observed a red light on the MSD side and realized the GC-MS was forced to shut down. We tried to conduct a pump down cycle and noticed the turbo pump was not able to start, the turbo pump speed was at 0% the whole time, after a few minutes the GC-MS shut down as a safety precaution. Every time we start the MS power, the red error shows right away.
We first checked if there is any major leaks around the GC and MS, it seems GC is fine, we plugged the column end to the MS transfer line to skip the GC part and nothing changed. Then we checked the vacuum knob and its O-ring, the side door of the EI source and closed them as rough pump started to run. It seems the rough pump works fine and the vacuum from the rough pump was pretty strong and normal. Our rough pump is a dry pump and we replaced the seal tip about 5 months ago and it has been running fine since then. To eliminate the rough pump problem, we switched to a different oil pump anyway and we still had the same problem.
On the instrument panel we observed an error message of “HIVAC Pump not Ready”, and the turbo pump speed was at 0% the whole time. We feel it should be something related to the turbo pump but we are not sure how to check on this part. Is there anything we can do to check this issue our own? Thank you so much!
Best regards,
Hill