Instrumentation information:
GC: 7890A on firmware A.01.04 (Currently only have back injector installed)
MS: 5975C on firmware 5.02.02
ALS: 7683B
Sample Tray: G2614A
PC: XP Professional running MSD ChemStation E.01.01.335
I'm trying to revive this instrument that hasn't been in use since 2019 and I'd just about gotten everything detected and set up but whenever I try to run a sequence through ChemStation, it goes through the entire process with the ALS just for it to inject the sample back into the sample bottle itself rather than into the injection port. I'm looking through documentation and it's difficult to tell if it's even being detected; I've looked through the msdchem.ini file and "ALS_ID= " is blank while "ALS_TYPE=0" leading me to believe I'm either missing something to get the GC to fully detect the ALS or there's something else wrong with it. Has anyone had any similar issues?