Aux EPC or PCM for Collision Gas 7890/7200

Hi, we're trying to set up a used 7200 and everything is working but the 7890A we have doesn't have the dedicated CC EPC - is it possible to configure a standard Aux EPC or PCM in a 7890A to control collision cell flow?

Thanks

Parents Reply Children
  •   - Will you actually be using IRM?  Very few use it on GC QTOF. The whole IRM setup is very touchy.

    US1247H001 was sold as an EI/CI mainframe, so the flow diagram Richard987 shared is correct. That means the CI subsystem flow controller provides the 0.5 ml/min of helium and not the CC EPC module. Make sure that there is never more than 30psi of any gas plumbed to the CI subsystem as higher pressures damage the flow controller. 

    The needle valve to the foreline pump sucks away much of the PFET as full headspace vapor out of the vial will overload the system.