Anybody have any advice on how to clean a HydroInert source? I only found suggestions to replace certain parts and not need to clean but that can be very costly.
Find maintenance, best practice, and troubleshooting articles.
Anybody have any advice on how to clean a HydroInert source? I only found suggestions to replace certain parts and not need to clean but that can be very costly.
When running hydrogen carrier your source should not need as much physical cleaning (compared to helium carrier) because the H2 has a scrubbing effect. So far I've been told that there is no physical cleaning on the HydroInert source as it could damage the inert coating. During testing our development team was able to run over 5,000 runs of extracted soil matrix before failing performance criteria. (For comparison, cleaning was required with helium carrier after less than 600 runs.) More details can be found at https://www.agilent.com/cs/library/technicaloverviews/public/te-hydroinert-source-5994-4889en-agilent.pdf
When running hydrogen carrier your source should not need as much physical cleaning (compared to helium carrier) because the H2 has a scrubbing effect. So far I've been told that there is no physical cleaning on the HydroInert source as it could damage the inert coating. During testing our development team was able to run over 5,000 runs of extracted soil matrix before failing performance criteria. (For comparison, cleaning was required with helium carrier after less than 600 runs.) More details can be found at https://www.agilent.com/cs/library/technicaloverviews/public/te-hydroinert-source-5994-4889en-agilent.pdf