Can I inject HS samples as splitless using the 7697A HS autosampler?

I have 6890N GC, 5975C MS, and 7697A HS Sampler and I am working with low sample volumes at a 0.1-0.01ng/mL concentration ranges. I was wondering if HS sampler can handle splitless injection.
Thank you for your help.

Parents
  • HS may handle splitless injection and maybe not , it is all according to your flow rates and pressures .

    in general when you want to apply splitless with HS you need to increase your column flow rate since HS flow maybe higher than your column so you system will not get ready .

    but on the other hand GC MS needs slow flow rates which doesn't make sense

    I recommend you first to adjust your ms method

    try low resolution , maybe you need to change ion source temp or you EMV .

    ...

    then try to optimize your head space method , maybe longer or shorter injction time or higher filling pressure .

    ...

    also try splitless injection , if the system get ready so try to inject and watch your repeatability , but make sure to increase your purge time to make sure that every thing make its way into column 

    good luck

  • Thank you, Mohammad! I'll work on improving the method in a way the splitless injection wouldn't be needed.

  • you are welcome , yes try the MS and HS method

    .....

    theory of incapability is simple

    Head space needs carrier gas flow to transfer sample from loop to inlet

    flow differs from method to another 

    but this flow is much higher than column flow 

    for example : column flow is 1.5 ml/min while HS transfer line flow 30 ml/min

    so splitless injection become impossible , and the instrument can't get ready status for injection

Reply
  • you are welcome , yes try the MS and HS method

    .....

    theory of incapability is simple

    Head space needs carrier gas flow to transfer sample from loop to inlet

    flow differs from method to another 

    but this flow is much higher than column flow 

    for example : column flow is 1.5 ml/min while HS transfer line flow 30 ml/min

    so splitless injection become impossible , and the instrument can't get ready status for injection