Can someone confirm the part number g3170-80103 is the correct part number for my 5975C detector? Thankyou in advance
Regards
Find maintenance, best practice, and troubleshooting articles.
Can someone confirm the part number g3170-80103 is the correct part number for my 5975C detector? Thankyou in advance
Regards
This would be correct for a Triple Axis Detector (TAD) . It is possible that you have a standard EM which would be 05971-80103. To confirm if you have a TAD, the MSD should have this noted on the covers of the MSD. You can also vent the instrument and the EM part number is noted on the HED assembly at the rear of the analyzer.
Regards
James
This would be correct for a Triple Axis Detector (TAD) . It is possible that you have a standard EM which would be 05971-80103. To confirm if you have a TAD, the MSD should have this noted on the covers of the MSD. You can also vent the instrument and the EM part number is noted on the HED assembly at the rear of the analyzer.