Headspace 7967A
Intuvo 9000 GC
5977B MSD
Has anyone had issues with the inlet liner cracking with the headspace assembly attached to the inlet?
Find maintenance, best practice, and troubleshooting articles.
Headspace 7967A
Intuvo 9000 GC
5977B MSD
Has anyone had issues with the inlet liner cracking with the headspace assembly attached to the inlet?
I'm also receiving a message error "Error code 20401000, cause: the EPC module is going into a shutdown, effect: the GC beeps to alert the user."
this error message occurs during my postrun.
I'm also receiving a message error "Error code 20401000, cause: the EPC module is going into a shutdown, effect: the GC beeps to alert the user."
this error message occurs during my postrun.
Make sure you are running SPLIT mode with a headspace injection and at least 20 ml/min of total flow. The GC electronic pressure control modules need at least 15 to 20 ml/min of total flow for reasonable control.
Do you have a Post Run time setup in your method? GC Oven page, Post Run Time.
If you have a post-run time there with a post run temperature, the system will also try to drive to the Post Run flow on the Columns page.