Time out reading APG controller, Chemstation GC7890A, MS5795C

-310,"({AIH}):Timeout waiting for APG RCI ready.",5201,"({AIH}):Time out reading APG controller interrupt cause"
and

-310 , "({SP00}MASK4;): APG RCI controller hardware not connected or not present."


I received the above mentioned messages multiple times during the last month. Is anyone familiar with this? it occurs during a sequence and stops the instrument.

 

We run GC 7890A with MSD 5795 C and the good ol' chemstation 02.02.1431

Parents
  • Hi,

     

    errors like this (APG) could caused by the autosampler and/or a broken or bad connected remote cable. Try to power off / restart your system and check/replace the cables.

    It could also causes by bad software and firmware, in this case upgrade FW to latest version or try to reinstall software.

    If nothing helps it could be also a bad communication board.

     

    BR

    pwd

Reply
  • Hi,

     

    errors like this (APG) could caused by the autosampler and/or a broken or bad connected remote cable. Try to power off / restart your system and check/replace the cables.

    It could also causes by bad software and firmware, in this case upgrade FW to latest version or try to reinstall software.

    If nothing helps it could be also a bad communication board.

     

    BR

    pwd