Autone problems 5975- peak width large

Autone problems 5975

 

I replace the electromultiplier due the lack of abundance (the filmanents are new too- all components from agilent), but I have problems in autone . problem reported " peaks width large". See the attach with parametrs tune with PFTBA close, when open ist still the same.

Can you help me? 

Parents
  • Your PFTBA valve is closed, set it to open and save the tune file, then try a tune again. It will close when you go back to the instrument screen.  If nothing changes then you have massive contamination from something in the MS or a board could be bad, you'll need a service call probably.

Reply
  • Your PFTBA valve is closed, set it to open and save the tune file, then try a tune again. It will close when you go back to the instrument screen.  If nothing changes then you have massive contamination from something in the MS or a board could be bad, you'll need a service call probably.