How do I change my autotune temperature settings?

I'm using a 7890/5975 with an inert ion source - my methods and manual tune utilize a source temperature above 300 degrees and I want to match my autotune to that as well. Currently if I want to autotune, my MS default temp will switch back to 250, but I want consistency between the two. 

Parents
  • Hi,

     

    you're able to set the temperatures in the tune and vacuum control menu. Go to PARAMETERS --> MANUAL TUNE, set the ion source temperature to 300 and save the tune file.

     

    If you tune the instrument, use TUNE --> TUNE MSD instead of ATUNE

     

    Warning: Your maximum allowed temperature depends on the type of your ion source, some of them have a maximum of 250 degrees.

     

    BR

    PWD

Reply
  • Hi,

     

    you're able to set the temperatures in the tune and vacuum control menu. Go to PARAMETERS --> MANUAL TUNE, set the ion source temperature to 300 and save the tune file.

     

    If you tune the instrument, use TUNE --> TUNE MSD instead of ATUNE

     

    Warning: Your maximum allowed temperature depends on the type of your ion source, some of them have a maximum of 250 degrees.

     

    BR

    PWD