I have several Agilent Mass Selective Detectors that reject heat. My room is overheating.
What is the air flow exhaust for:
5973
6890
7890A?
Find maintenance, best practice, and troubleshooting articles.
I have several Agilent Mass Selective Detectors that reject heat. My room is overheating.
What is the air flow exhaust for:
5973
6890
7890A?
Hi Buildingsystems,
The 5973 MSD is rated for 900watts(3000 BTU/hr)
The 6890: Standard Oven ramp: 7681 BTU/hr
Fast Oven ramp: 10,071 BTU/hr
The 7890: Same as 6890;
Hope this helps:
Mitch
Hi Buildingsystems,
The 5973 MSD is rated for 900watts(3000 BTU/hr)
The 6890: Standard Oven ramp: 7681 BTU/hr
Fast Oven ramp: 10,071 BTU/hr
The 7890: Same as 6890;
Hope this helps:
Mitch