High RSD with injections from same vial

Hi I am having precision issues with RSD of up 20% for injections from same vial for MDMA and diphenylamine. I am at method development stage and have changed septum, syringe and liner but have had no improvement. Can anyone suggest anything I can try?

  • Have you confirmed whether or not you get good precision from separate vials prepared with the same solution? If separate vials provide good precision, then I wouldn't think this issue is system related.

    Another thing to consider is whether there is a trend in a particular direction when injecting from the single vial. Do the injections consistently decrease/increase, or does it random? Trending in one direction would lead me to believe there is a stability issue with the sample in the punctured vial. I have had an assay where subsequent injections from a single vial are always lower, so in that case I prepare a separate vial for each injection.

  • Thanks Bryan I have tried it with the same solution in different vials and still have the same issue, there is no consistency it is fairly random

  • Under your method's Windows ***.D subdirectory is a file called "acqmeth.txt" that has all of the GC parameters, please share it.  What is the injection solvent? What sample prep?  Autosampler or manual injection? When you change the liner was it clean or were there little red particles of septum inside that result from an overtightened septum?  What is the RSD for a known standard like one of the instrument checkout samples using a known good method?   

    See: (+) Proving that an Agilent GC/MS system is fully functional - Files - GC/MS - Agilent Community

  • Hi Paul, it is an autosampler. The liner looks totally clean. Solvent is methanol.  I don't have a checkout sample, can you suggest anything else to use to assess precision? The analytes I am look at are MDMA and Diphenylamine. I have a feeling it might be method related as opposed to the instrument but I don't know how I can assess this as a new instrument and at method development stage so have nothing to compare it to  

    INSTRUMENT CONTROL PARAMETERS: GCMS-1
    ----------------------------------------

    D:\MassHunter\Methods\MDMA cunha 70_1.m
    Fri Jan 10 14:12:51 2025

    Control Information
    ------- -----------

    Sample Inlet : GC
    Injection Source : GC ALS
    Injection Location: Front
    Mass Spectrometer : Enabled


    No Sample Prep method has been assigned to this method.


    GC
    GC Summary
    Run Time 8 min
    Post Run Time 1 min

    Oven
    Temperature
    Setpoint On
    (Initial) 100 °C
    Hold Time 2 min
    Post Run 250 °C
    Program
    #1 Rate 20 °C/min
    #1 Value 200 °C
    #1 Hold Time 1 min


    Equilibration Time 1 min
    Max Temperature 320 °C
    Maximum Temperature Override Disabled
    Slow Fan Disabled

    ALS
    Front Injector
    Syringe Size 10 μL
    Syringe 10 μL
    Agilent 5181-3354
    Injection Volume 1 μL
    Solvent A Washes (PreInj) 3
    Solvent A Washes (PostInj) 0
    Solvent A Volume 8 μL
    Solvent B Washes (PreInj) 0
    Solvent B Washes (PostInj) 3
    Solvent B Volume 8 μL
    Sample Washes 0
    Sample Wash Volume 8 μL
    Sample Pumps 6
    Dwell Time (PreInj) 0 min
    Dwell Time (PostInj) 0 min
    Solvent Wash Draw Speed 300 μL/min
    Solvent Wash Dispense Speed 3000 μL/min
    Sample Wash Draw Speed 300 μL/min
    Sample Wash Dispense Speed 3000 μL/min
    Injection Dispense Speed 6000 μL/min
    Viscosity Delay 1 sec
    Sample Depth Disabled
    Injection Type Standard
    L1 Airgap 0.2 μL
    Solvent Wash Mode A, B

    Tray
    Tray Chiller/Heater
    Enable readiness check Off

    Barcode Symbology Enable all symbologies
    Barcode Checksum Off
    Vial Heater Disabled
    Vial Mixer Disabled

  •  Please reply with all the rest of the parameters listed in that file. It includes the inlet, oven, and MS, too.

  • Apologies Paul I thought I had copied it all, should all be there now

    INSTRUMENT CONTROL PARAMETERS: GCMS-1
    ----------------------------------------

    D:\MassHunter\Methods\MDMA cunha 70_1.m
    Fri Jan 10 14:12:51 2025

    Control Information
    ------- -----------

    Sample Inlet : GC
    Injection Source : GC ALS
    Injection Location: Front
    Mass Spectrometer : Enabled


    No Sample Prep method has been assigned to this method.


    GC
    GC Summary
    Run Time 8 min
    Post Run Time 1 min

    Oven
    Temperature
    Setpoint On
    (Initial) 100 °C
    Hold Time 2 min
    Post Run 250 °C
    Program
    #1 Rate 20 °C/min
    #1 Value 200 °C
    #1 Hold Time 1 min


    Equilibration Time 1 min
    Max Temperature 320 °C
    Maximum Temperature Override Disabled
    Slow Fan Disabled

    ALS
    Front Injector
    Syringe Size 10 μL
    Syringe 10 μL
    Agilent 5181-3354
    Injection Volume 1 μL
    Solvent A Washes (PreInj) 3
    Solvent A Washes (PostInj) 0
    Solvent A Volume 8 μL
    Solvent B Washes (PreInj) 0
    Solvent B Washes (PostInj) 3
    Solvent B Volume 8 μL
    Sample Washes 0
    Sample Wash Volume 8 μL
    Sample Pumps 6
    Dwell Time (PreInj) 0 min
    Dwell Time (PostInj) 0 min
    Solvent Wash Draw Speed 300 μL/min
    Solvent Wash Dispense Speed 3000 μL/min
    Sample Wash Draw Speed 300 μL/min
    Sample Wash Dispense Speed 3000 μL/min
    Injection Dispense Speed 6000 μL/min
    Viscosity Delay 1 sec
    Sample Depth Disabled
    Injection Type Standard
    L1 Airgap 0.2 μL
    Solvent Wash Mode A, B

    Tray
    Tray Chiller/Heater
    Enable readiness check Off

    Barcode Symbology Enable all symbologies
    Barcode Checksum Off
    Vial Heater Disabled
    Vial Mixer Disabled

    Sample Overlap
    Mode Sample overlap is not enabled

    ALS Errors Pause for user interaction

    Front SS Inlet He
    Mode Split
    Heater On 280 °C
    Pressure On 7.1442 psi
    Total Flow On 109.5 mL/min
    Septum Purge Flow On 3 mL/min
    Pre-Run Flow Test Off
    Gas Saver On 20 mL/min after 2 min
    Split Ratio 70 :1
    Split Flow 105 mL/min
    Liner Agilent 5190-2295: 870 μL (Split, taper, wool, low pressure drop)

    Column
    Column #1
    Flow
    Setpoint On
    (Initial) 1.5 mL/min
    Post Run 1 mL/min

    Column Information Agilent 122-5512UI
    Description DB-5ms Ultra Inert
    Temperature Range -60 °C—325 °C (350 °C)
    Dimensions 15 m x 250 μm x 0.25 μm (Uncalibrated)
    Heater Oven
    Column lock Unlocked
    In Front SS Inlet He
    Out MSD
    (Initial) 100 °C
    Pressure 7.1442 psi
    Flow 1.5 mL/min
    Average Velocity 64.594 cm/sec
    Holdup Time 0.38703 min
    Control Mode Constant Flow

    Column Outlet Pressure 0 psi

    Detector Evaluation
    Perform Detector Evaluation Test Off

    MSD Transfer Line
    Temperature
    Setpoint On
    (Initial) 280 °C


    Collision Cell/Aux
    Pressure
    Setpoint Off
    (Initial) 10 psi
    Post Run 10 psi

    Quench Gas He On 4 mL/min
    Collision Gas N2 On 1.5 mL/min
    Collision Cell Aux He (unused)

    PSD 1
    Pressure
    Setpoint Off
    (Initial) 4.3978 psi
    Post Run 60.198 psi

    PSD Purge Off
    PSD 1 He (unused)

    Signals
    Signal #1:
    Description None

    Signal #2:
    Description None

    Signal #3:
    Description None

    Signal #4:
    Description None

    Signal #5:
    Description None

    Signal #6:
    Description None

    Signal #7:
    Description None

    Signal #8:
    Description None


    END OF INSTRUMENT CONTROL PARAMETERS
    ------------------------------------