Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography
  • Technical Areas
  • More
Gas Chromatography
Wiki Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
  • Gas Chromatography
  • Agilent 8860, 8890, Intuvo 9000 GC Firmware
  • Agilent Intuvo 9000 GC Resources
  • Capillary column installation quick reference guide
  • Capillary Flow Technology - A Resource with Example Methods
  • Changing the Turret of the 7693A Automatic Liquid Sampler
  • Checking for External Leaks on 7697A Headspace
  • Collection of Agilent Micro GC Resources
  • Collection of GC Resources
  • Explore GC eLearning Courses at Agilent University
  • GC Calculators
  • Hardware reset for 6890, 7890 and 6850
  • How to Align the Agilent 7683B GC Injector
  • How to align the Agilent 7693A GC injector
  • How to Calibrate the Agilent 7693A Sample Tray
  • How to turn LED light on and off on 7693 ALS
  • Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890
  • Removing and Installing a Syringe on the Agilent 7693A Autosampler
  • Replacing the Syringe Carriage Assembly in the Agilent 7693A Automatic Liquid Sampler
  • Troubleshooting an FID not igniting
  • Turret changing procedure for 7683B autosampler

Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890

The following are the manual steps for performing the standard inlet tests for 6890 and 7890A GCs. Typically, one runs the Inlet Leak Check to confirm whether there's a leak (the leak rate you determine is only approximate), the symptom of which is usually low pressure and/or smaller, longer retention time peaks; if there is a leak confirmed, one would follow this test up with the Pressure Decay test, which is more stringent and precise. The Split Vent Restriction Test addresses the opposite problem from a leak: a constriction, which typically causes high pressure.

 

6890 Inlet Leak Check (can also be used for 7890)

  1. Set the mode to Splitless.
  2. Confirm column dimensions – leave the column installed in the inlet
  3. Set the Column Flow to 4 mL/min.
  4. Set the inlet temperature to a constant value.
  5. Set the inlet purge flow to 50 mL/min.
  6. Press the Prep Run button once, wait a few seconds, then press Prep Run again.
  7. Monitor the Total flow.

Subtract 7mL/min from the total flow.  The 7mL/min is the column flow + septum purge (4mL/min + 3mL/min).  The number you get quantifies the approximate leak rate; obviously, this number is ideally close to or equal to 0.  To fix the inlet leak or to identify its cause, check/tighten the following while monitoring the total flow: septum, liner O-ring, gold seal,  column fitting, split vent trap and associated fittings, and gang block fittings.

 

6890 Inlet Pressure Decay Test

This test requires capping the inlet column and septum purge fittings. The inlet is pressurized, then the pressure is shut off. The leak rate is then indicated by the pressure decay.

  1. Set the GC inlet to split mode.
  2. Remove the column and cap the inlet fitting.
  3. Cap the septum purge fitting.
  4. Set the inlet temperature to a constant temperature.
  5. Set inlet pressure to 25 PSI.
  6. Set the total flow to 50 mL/min.
  7. Allow one minute for the pressures to stabilize.
  8. Set the pressure to OFF.
  9. Turn off the carrier supply to the inlet
  10. Monitor the pressure for 10 minutes.

The pressure should not drop by more than 0.5 PSI in that time period.  If it does, there’s likely a leak. Consider the inlet body itself if the consumables were replaced when the Inlet Leak Check was performed.

 

7890 Inlet Pressure Decay Test

This test requires capping the inlet column and septum purge fittings. The inlet is pressurized, then the pressure is shut off. The leak rate is then indicated by the pressure decay.

  1. Set the GC inlet to split mode.
  2. Remove the column and cap the inlet fitting.
  3. Cap the septum purge fitting.
  4. Set the Septum Purge Valve Duty Cycle to 50%.

          Press the blue Service Mode > Diagnostics > Front or Back inlet (as appropriate for what you are testing) > Pneumatics control > Septum Purge > Duty Cycle 50%.

  1. Set the inlet temperature to a constant temperature.
  2. Set inlet pressure to 25 PSI.
  3. Set the total flow to 60 mL/min.
  4. Allow one minute for the pressures to stabilize.
  5. Set the pressure to OFF.
  6. Turn off the carrier supply to the inlet
  7. Monitor the pressure for 10 minutes.

The pressure should not drop by more than 0.5 PSI in that time period. If it does, there’s likely a leak. Consider the inlet body itself if the consumables were replaced when the Inlet Leak Check was performed.

 

6890/7890 Split Vent Restriction Test

  1. Set the mode to split
  2. You can either have the column installed or the inlet capped.
  3. De-configure the column from the inlet.
  4. Set the inlet temperature to a constant value.
  5. Set the inlet pressure to 0 PSI (this is not the same as turning the pressure OFF).
  6. Set the inlet total flow to 400 ml/minute.
  7. Monitor inlet actual pressure.  Typical Values for a new, clean GC are, for split liner 1-2 PSI, and for a splitless liner, 3-10 PSI.  Much higher values indicate condensed sample contamination in the injection port split vent tube, the copper split vent line or the split vent trap, or possibly a defective EPC.
  • troubleshooting
  • agilent GC Instruments
  • maintenance
  • Share
  • History
  • More
  • Cancel
Anonymous
Parents
  • topsoe-lak
    topsoe-lak over 2 years ago

    Hello,

     

    Are the flow and pressure settings for the manual leak tests for helium as carrier gas only or will it work also if hydrogen or nitrogen is used as carrier gas?

     

    With kind regards -

    Lars

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
  • valentinrusu
    valentinrusu over 2 years ago in reply to topsoe-lak

    They should be applicable to any carrier gas, as long as what is being used physically is also configured on the GC. The inlet EPC module controls pressure based on the properties of the gas type configured. 

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
Comment
  • valentinrusu
    valentinrusu over 2 years ago in reply to topsoe-lak

    They should be applicable to any carrier gas, as long as what is being used physically is also configured on the GC. The inlet EPC module controls pressure based on the properties of the gas type configured. 

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
Children
No Data

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help