Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Gas Chromatography
  • Technical Areas
Gas Chromatography
Wiki Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890
  • Announcements
  • Blog
  • Forum
  • Articles
  • Videos
  • Wiki
  • More
  • Cancel
Actions
  • Share
  • More
  • Cancel
Table of Contents
  • +Collection of Agilent GC Resources Home
  • Capillary Flow Technology - A Resource with Example Methods
  • Chromatography with Supercritical Fluids SFC and Comparison with GC and HPLC
  • Explore GC eLearning Courses Available from Agilent Education
  • Firmware for Agilent GCs, Headspace Samplers, Autosamplers and More
  • How to Access the Agilent GC Calculators and Method Translation Software
  • Hardware reset for 6890, 7890 and 6850
  • How to turn LED light on and off on 7693 ALS
  • Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890
  • Peak Tales HPLC and GC Chromatography Podcast
  • Turret changing procedure for 7683B autosampler
Still Need Help?

Post your question in our User Forum or Contact Support.

Manual Steps for Inlet Leak Check, Pressure Decay Test, and Split Vent Restriction Test for 6890/7890

Created by valentinrusu valentinrusu over 2 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 1 year ago

The following are the manual steps for performing the standard inlet tests for 6890 and 7890A GCs. Typically, one runs the Inlet Leak Check to confirm whether there's a leak (the leak rate you determine is only approximate), the symptom of which is usually low pressure and/or smaller, longer retention time peaks; if there is a leak confirmed, one would follow this test up with the Pressure Decay test, which is more stringent and precise. The Split Vent Restriction Test addresses the opposite problem from a leak: a constriction, which typically causes high pressure.

 

6890 Inlet Leak Check (can also be used for 7890)

  1. Set the mode to Splitless.
  2. Confirm column dimensions – leave the column installed in the inlet
  3. Set the Column Flow to 4 mL/min.
  4. Set the inlet temperature to a constant value.
  5. Set the inlet purge flow to 50 mL/min.
  6. Press the Prep Run button once, wait a few seconds, then press Prep Run again.
  7. Monitor the Total flow.

Subtract 7mL/min from the total flow.  The 7mL/min is the column flow + septum purge (4mL/min + 3mL/min).  The number you get quantifies the approximate leak rate; obviously, this number is ideally close to or equal to 0.  To fix the inlet leak or to identify its cause, check/tighten the following while monitoring the total flow: septum, liner O-ring, gold seal,  column fitting, split vent trap and associated fittings, and gang block fittings.

 

6890 Inlet Pressure Decay Test

This test requires capping the inlet column and septum purge fittings. The inlet is pressurized, then the pressure is shut off. The leak rate is then indicated by the pressure decay.

  1. Set the GC inlet to split mode.
  2. Remove the column and cap the inlet fitting.
  3. Cap the septum purge fitting.
  4. Set the inlet temperature to a constant temperature.
  5. Set inlet pressure to 25 PSI.
  6. Set the total flow to 50 mL/min.
  7. Allow one minute for the pressures to stabilize.
  8. Set the pressure to OFF.
  9. Turn off the carrier supply to the inlet
  10. Monitor the pressure for 10 minutes.

The pressure should not drop by more than 0.5 PSI in that time period.  If it does, there’s likely a leak. Consider the inlet body itself if the consumables were replaced when the Inlet Leak Check was performed.

 

7890 Inlet Pressure Decay Test

This test requires capping the inlet column and septum purge fittings. The inlet is pressurized, then the pressure is shut off. The leak rate is then indicated by the pressure decay.

  1. Set the GC inlet to split mode.
  2. Remove the column and cap the inlet fitting.
  3. Cap the septum purge fitting.
  4. Set the Septum Purge Valve Duty Cycle to 50%.

          Press the blue Service Mode > Diagnostics > Front or Back inlet (as appropriate for what you are testing) > Pneumatics control > Septum Purge > Duty Cycle 50%.

  1. Set the inlet temperature to a constant temperature.
  2. Set inlet pressure to 25 PSI.
  3. Set the total flow to 60 mL/min.
  4. Allow one minute for the pressures to stabilize.
  5. Set the pressure to OFF.
  6. Turn off the carrier supply to the inlet
  7. Monitor the pressure for 10 minutes.

The pressure should not drop by more than 0.5 PSI in that time period. If it does, there’s likely a leak. Consider the inlet body itself if the consumables were replaced when the Inlet Leak Check was performed.

 

6890/7890 Split Vent Restriction Test

  1. Set the mode to split
  2. You can either have the column installed or the inlet capped.
  3. De-configure the column from the inlet.
  4. Set the inlet temperature to a constant value.
  5. Set the inlet pressure to 0 PSI (this is not the same as turning the pressure OFF).
  6. Set the inlet total flow to 400 ml/minute.
  7. Monitor inlet actual pressure.  Typical Values for a new, clean GC are, for split liner 1-2 PSI, and for a splitless liner, 3-10 PSI.  Much higher values indicate condensed sample contamination in the injection port split vent tube, the copper split vent line or the split vent trap, or possibly a defective EPC.
  • Troubleshooting
  • Maintenance
  • GC
  • Share
  • History
  • More
  • Cancel
  • Sign in to reply
Parents
  • AMatta
    AMatta over 2 years ago

    Are the settings below suitable for the 8860 model?

    1. Set the inlet temperature to a constant temperature.
    2. Set inlet pressure to 25 PSI.
    3. Set the total flow to 60 mL/min.
    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Comment
  • AMatta
    AMatta over 2 years ago

    Are the settings below suitable for the 8860 model?

    1. Set the inlet temperature to a constant temperature.
    2. Set inlet pressure to 25 PSI.
    3. Set the total flow to 60 mL/min.
    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Children
No Data
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Privacy Statement
Terms of Use
Contact Us
Site Help