Hello, I am looking for some answers on high reaction cell pressure. I performed reaction cell cleaning on 8355 SCD just for maintenance purposes. Afterwards, reaction cell pressure would not pull down past 18 Torr. I followed start-up protocols and double-checked all fittings and O-ring placements. I capped off the transfer line and the issue is isolated to the SCD. Oddly, removing the ozone scrubber/trap lets the reaction cell pull down immediately to 3 torr. I tried several new ozone traps with the same 18 torr reaction cell pressure. Before I performed the cell cleaning the reaction cell pressure was 6 torr. What am i missing?