  • Q: Agilent7890B. During the injection. the ALS clip dropped and the ALS cannot move. The ALS reported an 220 error. And it is displayed that the sample is being prepared.


    A: If you mean the black clip is dropped and can not install on. It need a repair by engineer. error 220 is a front Injector Turret Axis Errors. Most happened when the turret is stuck.You can disconnect the power supply and turn the turret manually. If it turns smoothly, restart the power self-test.
    You seems met 2 separate problems. If these move cannot fix the problems. Please contact your local Agilent Service center.

    For local support in China, you can contact us via Agilent China's service hotline 800-820-3278, mobile phone 400-820-3278. Or consult related questions through Agilent's WeChat applet by scan the QR code below.