Hi all ! There was such an error when entering a sample (on the picture), nowhere found a description of this error and how to fix it. Does anyone know what could be the cause? Thank you
Find maintenance, best practice, and troubleshooting articles.
Hi all ! There was such an error when entering a sample (on the picture), nowhere found a description of this error and how to fix it. Does anyone know what could be the cause? Thank you
The problem was solved by flashing the board on the GC itself, now everything works. I assume the electronic module that is in the GC itself and responsible for communication with the auto sampler began to malfunction.
Hi , can you please tell me if flashing of the als firmware on the gc controller was done or flashing of the GC firmware was done ?