SCD Burner Heating

We use 8355 SCD Detector. We always face issues with the burner heating. The set temp is 800 C and the actual value always show a fixed value of 1263 C. Is it the problem with the thermocouple? When we try to change the position of the thermocouple inside the burner, the temperature varies. Kindly advise

  • It sounds like a bad connection on the thermocouple or a problem with the thermocouple itself.  If the thermocouple is securely plugged into the back of the 8355, I recommend scheduling a visit by an Agilent service representative.  Contact phone numbers for your area can be found on our webpage.  You can also log a service request online on that page - click on "Request Technical Support."