The LED light is ON by default, but I can't find a way to turn it off, either from the GC control panel or from the web browser. Please advise as necessary.
Find maintenance, best practice, and troubleshooting articles.
The LED light is ON by default, but I can't find a way to turn it off, either from the GC control panel or from the web browser. Please advise as necessary.
Hi raphasanz
You should be able to turn the ALS light off from the Intuvo Touchscreen.
1. Press Settings on the touch screen control ribbon. The Settings view appears.
2. Press Configuration.
3. You should have an ALS/Tray menu item, uncheck Injector Light On (refer to the screen capture below).
Refer to this procedure and other Configuration Settings in the Intuvo 9000 Gas Chromatograph Operation Manual (agilent.com)
Cheers!
Greatly appreciated!