Hello everyone, in G4513A There is a black rubber band on the gear that starts to tear and cause a malfunction "plunger error". Is there a service note about that? Is there a place where it is writen about that? I cant find it. Thanks
Hello everyone, in G4513A There is a black rubber band on the gear that starts to tear and cause a malfunction "plunger error". Is there a service note about that? Is there a place where it is writen about that? I cant find it. Thanks
Hi,
I replaced the band with a black sealing ring of about the same size. The band is probably only necessary to hold on the set position.
My autosampler is running fine again. Cost under 1 € but there is no warranty
Best regards