Hello everyone, in G4513A There is a black rubber band on the gear that starts to tear and cause a malfunction "plunger error". Is there a service note about that? Is there a place where it is writen about that? I cant find it. Thanks
Hi
There is no service note available online.
U have to send autoinjector to Agilent for repair to resolve the issue of plunger error.
Sometime plunger motor, sometime belt, sometime syringe carrier assy are faulty so need service call for autoinjector...
Hello,
I found a service note about this error, there it is writen:
Replace the whole syringe carriage when syringe carriage O ring of G4513A/G4567A is broken or deteriorated Summary: Replace the whole syringe carriage G4513-67970 when syringe carriage O ring of G4513A/G4567A is broken or deteriorated after several years’ usage or by some solvents used.
Thank you ISH,
I appreciate it.