Transitioning from Packed to Capillary Column (GC 7890B)

Does anyone have experience transitioning column types from packed to capillary on the GC? Our current inlet is purged packed and solvents are polar, so inlet volume is a concern. Any tips or recommendations on how best to handle this type of transition would be very appreciated. Thanks!  

Parents Reply Children
  • And detector side, need to install capillary adaptor...

  • We won't know if a new inlet is needed until trying (with our solvents we are borderline for volume). From what we've been told we may not need a new inlet. Do you have any experience in how the operational procedure changes between a packed and capillary column? 

  • There are big difference..

    Major is packed column use high flow like 10/15/30ml while capillary column used to maximum 5ml of flow generally.

    Configuration is totally different..

    By the way, why u want to change it...Any specific application requirement?

  • Find above detail of all type agilent inlets and there use and differences...

  • Which packed column and detector are you using?  What volume are you injecting? What detection limit do you need?