8355 SCD high output

I operate an Agilent 8355 SCD. The  baseline output that is stated in the manual is 150pA in a well stabilized system with an oven temperature of 50C. When not in use, the baseline output is around 500pA. I'm not sure if this is caused by the method I use/frequency of use or if there is an issue with the detector itself that needs to be addressed. Any thoughts on the cause of a high baseline output?

Parents
  • The SCD is fixed. The reaction cell was replaced and the tech adjusted the oxidizer/hydrogen flow ratio. The oxidizer flow was lower than the hydrogen flow which can cause problems over time. The output is currently fluctuating between 50 and 60 pA while sitting. 

Reply
  • The SCD is fixed. The reaction cell was replaced and the tech adjusted the oxidizer/hydrogen flow ratio. The oxidizer flow was lower than the hydrogen flow which can cause problems over time. The output is currently fluctuating between 50 and 60 pA while sitting. 