I have a 7890B GC. Can I fit a packed column to this GC. I am trying to test for Acrylonitrile at the ppm levels with this GC.
Find maintenance, best practice, and troubleshooting articles.
I have a 7890B GC. Can I fit a packed column to this GC. I am trying to test for Acrylonitrile at the ppm levels with this GC.
If you have Purged Packed Inlet in GC than can fit packed column to this GC .
A 7890B is capable of supporting packed columns, but you will need a Purged Packed Inlet on the GC.
Inlet should be modified to purge packed inlet and detector fitting for column should be modified for packed column installation..