Hello,
I would like to know how many injections were made on a 7820A Agilent GC between 2 dates.
Is there a way through the software (Openlab CDS Chemstation GC Agilent) to find this information?
Thank you.
Find maintenance, best practice, and troubleshooting articles.
Hello,
I would like to know how many injections were made on a 7820A Agilent GC between 2 dates.
Is there a way through the software (Openlab CDS Chemstation GC Agilent) to find this information?
Thank you.
Greetings,
instead of EMF, is it impossible to get this information from older instrumentation. Software does not have this functionality.
As workaround you can export Activity Log from Control Panel and simply filter it for "Injection". You can get number of injections between two dates after some suitable fitlers.