So the GC is under Empower control, gets ready to inject and then does not inject, throwing this error into message center:
Not a valid back injector range. Back injector range is between 501-999.
Any ideas??
Find maintenance, best practice, and troubleshooting articles.
So the GC is under Empower control, gets ready to inject and then does not inject, throwing this error into message center:
Not a valid back injector range. Back injector range is between 501-999.
Any ideas??
Turns out that when using the rear port, vial numbering has to begin with 501. Do not ask why, I do not know.
Turns out that when using the rear port, vial numbering has to begin with 501. Do not ask why, I do not know.
What vail number is used if the head space is connected to the front port?