8355 SCD detector low response

I have my GC/SCD for analyzing butane samples. After I did maintenance on my pump for the SCD, I lost my signal. All I did was change and top off the oil. I was able to get a little response but the repeatability is not good. I use to get around 7000 area counts with my 10 ppm isopropyl mercaptan standard, but now I only get 2000. I tried changing both inner and outer tubes. I checked the reaction cell and the flow in the restrictor and it was good. I also just use oxygen for the ozone generator. Is there a recommended way for doing maintenance? Do I have to change everything when I do any maintenance? 

Parents
  • Hello,

     

    pump oil mist filter:
    the pump used to be equipped with 2 filters.
    * The long l-shaped cylinder = ozone barrier (this should be changed regularly).
    * (experience on GC MS) the cubic oilmist filter can be overlooked as a PM part & the oil separator can soak up with oil if excessive gas ballast is used.
    For simple / short-term tests: remove the cube-shaped Oilmist filter & apply a suction hood to the pump.
    a) does the vacuum change?
    b) is the SCD sensitivity better?


    * vacuum pressure value: a very important value that I wrote down regularly. When the value has been determined, a sticker on the display can be helpful (then everyone should remember to always look at it), because poor vacuum explains a problem with the sensitivity.

     

    * the drop test: if possible do not let anything drip on it, so that there is no later overload of the SCD.

    It can be enough to open a small "smelly" sulfur vial & hold it on all screw connections, since many screw connections are under vacuum.

     

    * tighten / swaging all connectors:
    (Unless otherwise specified in the manual) Do not overtighten the connectors. Hand tight & maximum one 1/16 (- 1/8) turn. When the burner got hot, some seals may still be leaking; But the same applies here: gentle tightening (1/16 turn).
    But before I tight seals I always observed the vaccum value and used the "stinking" vial / drop.

     

    * wiping cleaning:
    sorry I wasn't clear: the PMT is only being removed & not cleaned! Cleaning is OK for: buner cell & filter disc with lint-free cloth & solvent.

     

    * ceramic tubes
    even if these parts cost something, I always swapped them in pairs, because the time until everything was conditioned took too long ;-)

     

    I cannot give taht many general recommendations because it was all 20 years ago.


    I suggest: asking Agilent here again what there is to be learned from experience/empirical values.
    best might be: based on these current tests and our own experience write a new message here with final question & actual SCD status.

     

    regards,

Reply
  • Hello,

     

    pump oil mist filter:
    the pump used to be equipped with 2 filters.
    * The long l-shaped cylinder = ozone barrier (this should be changed regularly).
    * (experience on GC MS) the cubic oilmist filter can be overlooked as a PM part & the oil separator can soak up with oil if excessive gas ballast is used.
    For simple / short-term tests: remove the cube-shaped Oilmist filter & apply a suction hood to the pump.
    a) does the vacuum change?
    b) is the SCD sensitivity better?


    * vacuum pressure value: a very important value that I wrote down regularly. When the value has been determined, a sticker on the display can be helpful (then everyone should remember to always look at it), because poor vacuum explains a problem with the sensitivity.

     

    * the drop test: if possible do not let anything drip on it, so that there is no later overload of the SCD.

    It can be enough to open a small "smelly" sulfur vial & hold it on all screw connections, since many screw connections are under vacuum.

     

    * tighten / swaging all connectors:
    (Unless otherwise specified in the manual) Do not overtighten the connectors. Hand tight & maximum one 1/16 (- 1/8) turn. When the burner got hot, some seals may still be leaking; But the same applies here: gentle tightening (1/16 turn).
    But before I tight seals I always observed the vaccum value and used the "stinking" vial / drop.

     

    * wiping cleaning:
    sorry I wasn't clear: the PMT is only being removed & not cleaned! Cleaning is OK for: buner cell & filter disc with lint-free cloth & solvent.

     

    * ceramic tubes
    even if these parts cost something, I always swapped them in pairs, because the time until everything was conditioned took too long ;-)

     

    I cannot give taht many general recommendations because it was all 20 years ago.


    I suggest: asking Agilent here again what there is to be learned from experience/empirical values.
    best might be: based on these current tests and our own experience write a new message here with final question & actual SCD status.

     

    regards,