when we are trying injec VOC, this image appear and the frase in system says: "Syringe volume out of range"
when we are trying injec VOC, this image appear and the frase in system says: "Syringe volume out of range"
Hello James.
Sure...
Software: MSD Chem Station
GC: 7890A - Serial number: CN12321047
GC MS: 5975C with triplex-Axis detector - Serial Number: US12342A07
Autosampler: GC sampler 80 - Serial number: G6501B
Thank you
Regards
Camila Lima
The method is old, we used it for a long time, until last week everything was ok, however, since Friday it started showing such errors.
Syringe size configured for sampler is: 2.5 mL CTC Headspace, 23g - Agilent
volume of sample used for injection: 1 mL
Thank you
Yes, We have.
The code of syringe is ID28.
Hi camilalima,
Can you let us know what software and the model of GC and autosampler you are using?
Regards
James
Thanks Camila,
What syringe size do you have configured on the Sampler 80? Is it a new method? If so is the problem only with this method?
Regards
James
Have you verified the syringe holder is for the headspace syringe?
We have gotten a similar error. Turns out our syringe was pretty dirty. We ended up swapping for a different one and we having seen it since.
Hello,
Have you rebuilt the method from the default method? This would rule out corruption as the cause of the issue. The error code refers to a plunger position error. This could be the result of a dirty syringe or a hardware issue with the injection unit or you are trying to inject too fast. Given that this method has run for awhile I would check the syringe or you could possibly need service on your injection unit.
Hi camilalima,
Were you able to resolve the issue?
Regards
James
Hi James.
We checked the software settings and noticed something wrong, when we fixed the software link, it went back to normal, we deleted the access link and configured it again.