Can anyone advise me on how to control an external valve and what cable to use on a 6890N

I just inherited a 6890N GC. I what to put on an external Gas sampling valve. Can anyone advise me on how to control the valve thru mass hunter and how what cable to use to control the valve. It is an air operated 2 position valve

Parents
  • Hello  ,

    There are a total of 8 valve positions on your inherited GC. To keep things simple, valves 7-8 are the valve positions you are looking for, they can trigger external valves, contact closure, 48 vdc or 48 vac max. Using the contact closure is how you would trigger external devices.

    You would need to use cable PN G1530-60590 for this.

    If memory serves, In the method editor in Masshunter you have an Events section where you can trigger valve On/Off.

    I hope this helps.

Reply
  • Hello  ,

    There are a total of 8 valve positions on your inherited GC. To keep things simple, valves 7-8 are the valve positions you are looking for, they can trigger external valves, contact closure, 48 vdc or 48 vac max. Using the contact closure is how you would trigger external devices.

    You would need to use cable PN G1530-60590 for this.

    If memory serves, In the method editor in Masshunter you have an Events section where you can trigger valve On/Off.

    I hope this helps.