Volume Accuracy Failure - 850-DS Dissolution Sampling

During volume accuracy test of our 850-DS Dissolution Sampling, the difference of the sample volume for position #2 for 10mL was higher than 0.5mL, being that specification is lower than 0.25mL.

How can we calibrate the equipment to sample the right volume?

We performed the full range volume dispense (in the calibration screen), but we are not sure how to use the data to do the equipment calibration.

Could you help us, please?

# Empty Test Tube Filled Test Tube Difference Actual Sample Volume Difference from Set Point of 10 mL
1 7.6591 17.6115 9.9524 10.0226 -0.0226
2 7.6452 17.0647 9.4195 9.4859 0.5141
3 7.5446 17.4167 9.8721 9.9417 0.0583
4 7.6616 17.4081 9.7465 9.8152 0.1848
5 7.6411 17.5016 9.8605 9.9300 0.0700
6 7.6298 17.4850 9.8552 9.9247 0.0753
7 7.5995 17.4570 9.8575 9.9270 0.0730
8 7.5819 17.4926 9.9107 9.9806 0.0194

Thank you so much.

Suzany

Parents
  • Hello Suzany-
    Before do volume accuracy, one should perform volume calibration and it can be done using the calibration Tab key. 
    Full Procedure can be found on your initial OQ or Lycrecia provided manual. 
    Position 2 seems way off; you might need to contact Agilent service. 

    Thanks
    Kind Regards
    Khurrum Dar
    Field System Engineer
    Agilent Technologies, Life Sciences & Chemical Analysis
    Phone : 845-608-3376
    E-Mail : khurrum.dar@agilent.com

    You cannot live a perfect day without doing something for someone who will never be able to repay you.
Reply
  • Hello Suzany-
    Before do volume accuracy, one should perform volume calibration and it can be done using the calibration Tab key. 
    Full Procedure can be found on your initial OQ or Lycrecia provided manual. 
    Position 2 seems way off; you might need to contact Agilent service. 

    Thanks
    Kind Regards
    Khurrum Dar
    Field System Engineer
    Agilent Technologies, Life Sciences & Chemical Analysis
    Phone : 845-608-3376
    E-Mail : khurrum.dar@agilent.com

    You cannot live a perfect day without doing something for someone who will never be able to repay you.
Children
Was this helpful?