Here is a list of the currently available Knowledgebase for Agilent's consumable products in the Chemistries and Supplies Division (CSD) in the following categories:
Filaments and Ion Sources
Question Q001 What are the steps to increase filament life? Answer A001 https://www.agilent.com/en-us/support/mass-spectrometry/kb001483 Question Q002 What are the steps to troubleshoot a broken filament? Answer A002 https://www.agilent.com/en-us/support/mass-spectrometry/kb001602 Question Q003 What filaments are supported on the MSD? Answer A003 https://www.agilent.com/en-us/support/mass-spectrometry/kb001503 Question Q004 What is the procedure to maintain the MS Filaments? Answer A004 https://www.agilent.com/en-us/support/mass-spectrometry/kb001604
Gas Management
Question Q005 Are there any concerns about the temperature of the gas or the surroundings when using the Agilent G6691A Flowmeter?? Answer A005 The gas flow can be any temperature. The storage temperature for the Flow Meter will affect accuracy. The unit should be stored at temperatures under 50 degrees Celsius. A sticker is attached to each unit to gauge if the Flow Meter has gone outside the recommended temperature range while in storage or transit. If that happens, the sticker on the unit will turn black. The Flow Meter should then be sent back as a DEFOA and a replacement sent to the customer. Question Q006 Can I recycle my gas traps? Answer A006 Generally no, accept for Renewable Gas Purification Systems http://www.agilent.com/en-us/products/lab-supplies/chromatography-spectroscopy/gas-management/renewable-gas-purification-system Question Q007 Can I use rechargeable batteries in the Agilent G6691A Flowmeter? Answer A007 Yes, but the Flow Meter won’t read full battery power; cell voltage with the rechargeable batteries will be less than with alkaline batteries. Question Q008 Can it measure corrosive gases (e.g. ammonia)? Answer A008 No. This is not approved by Agilent. If measured, the warranty is invalidated and Agilent can’t guarantee the future performance or accuracy of the Flow Meter. The Flow Meter is designed for dry, noncorrosive gases. Question Q009 Can the Agilent G6691A Flowmeter measure corrosive gases (e.g. ammonia)? Answer A009 No. This is not approved by Agilent. If measured, the warranty is invalidated and Agilent can’t guarantee the future performance or accuracy of the Flow Meter. The Flow Meter is designed for dry, noncorrosive gases. Question Q010 Does the cartridge for the Agilent G6691A Flowmeter have a shelf life? Answer A010 The recommended shelf life is one year plus one additional year of use. The box it ships in has a temperature sticker stating if the cartridge is stored at a temperature above 130 F that could damage the components. Question Q011 Does the temperature of the gas affect the measurement on the Agilent G6691A Flowmeter? Answer A011 Yes. For a given mass flow rate the volumetric flow rate of a gas vary with the gas temperature and pressure. For a given mass flow rate the volumetric flow rate will increase with an increase in gas temperature and a decrease in gas pressure. As the ADM Flow Meter is a volumetric flow meter it will report the flow rate as it is measured. In mass flow mode the temperature and pressure are taken into account and the flow rate is displayed as if the gas temperature was 0 °C and 1 atm of pressure. Question Q012 How do I download the calibration certificate for the Agilent G6691A Flowmeter? Answer A012 Consult the Flow Meter Operating Instructions https://www.agilent.com/cs/library/usermanuals/public/G6691-90000.pdf Question Q013 How do I get a replacement hose for the Agilent G6691A Flowmeter? Answer A013 The part number for the replacement hose is G6692-68000. Question Q014 How do I know when to change my GasClean Filers? Answer A014 https://www.agilent.com/cs/library/usermanuals/public/5973-1528_GasCleanFilter_UserManual_EN.pdf Question Q015 How do I update the firmware on the Agilent G6691A Flowmeter? Answer A015 Consult the Flow Meter Operating Instructions https://www.agilent.com/cs/library/usermanuals/public/G6691-90000.pdf Question Q016 How many points are on the Certificate of Calibration for the Agilent G6691A Flowmeter? Answer A016 The unit is calibrated using 12 points and verified at 14 flow rates, which are show on the certificate. The verification points are: 0,0.5,1.0,5.0,10,50,100,300,400,500,601,700,750 ml/min. Question Q017 I have a Agilent G6691A Flowmeter. What do I do with the spare cartridge after it is out of calibration? Answer A017 The consumed cartridge can be disposed of per your local requirements for electronic waste. A new cartridge can be ordered online at https://www.agilent.com/en-us/products/lab-supplies/chromatography-spectroscopy/gas-management/adm-flow-meter#buy-products Question Q018 I have a Gas Clean filter, what does it mean if the indicator is changing colors from the bottom up? Top down? Answer A018 https://www.agilent.com/cs/library/usermanuals/public/5973-1528_GasCleanFilter_UserManual_EN.pdf Question Q019 If I cut the hose on the Agilent G6691A Flowmeter?, will that affect my measurements? Answer A019 Changing the hose length will affect your measurements. The Flow Meter is calibrated with the hose. Question Q020 Is it safe to measure hydrogen with the Agilent G6691A Flowmeter? Answer A020 Yes. Agilent advises to purge with an inert gas after measuring hydrogen. Question Q021 Is the calibration start date for the Agilent G6691A Flowmeter what is listed electronically inside the flowmeter or is it what is listed on the calibration certificate shipped with the flow module? How do we justify this with our auditors when they ask Answer A021 It is what is listed electronically inside of the flowmeter. The flow cell calibration is valid for one year after installation. If the customer is unable to follow this guideline due to their local procedures, they will need to replace the cartridge based on the calibration certificate date rather than the Agilent recommended date provided on the screen. Question Q022 Is there a user manual supplied with the Agilent G6691A Flowmeter? Answer A022 The only documentation provided with the flowmeter is the calibration certificate and the Quick Start Guide. The user manual can be found on the product page on www.Agilent.com Question Q023 Is there an AC Adapter for the Agilent G6691A Flowmeter? Answer A023 You can use the USB cable but will need to purchase the appropriate plug for the cord. The unit can be powered by the USB, but it will not recharge the batteries. Question Q024 Is there instructions on how a measurement is done for split ratio using the Agilent G6691A Flowmeter? Answer A024 We developed a detailed Troubleshooting Guide that details this measurement at https://www.agilent.com/cs/library/technicaloverviews/public/5991-7685EN.pdf Question Q025 Is there instructions on how a measurement is done for split ratio? Answer A025 We developed a detailed Troubleshooting Guide that details this measurement at https://www.agilent.com/cs/library/technicaloverviews/public/5991-7685EN.pdf Question Q026 My Flow Meter OLED screen reads “Error code X.” What does that mean? Answer A026 All error codes are listed in the Operation Manual for the Flow Meter. Question Q027 The Agilent G6691A Flowmeter is supposedly able to be powered by both batteries as well as through USB. Why doesn’t the USB cable come with a power adaptor? Answer A027 The USB adapter is intended for connecting to the PC for updating the firmware or accessing the digital version of the calibration certificate. The Flow Meter is powered by 3 AA batteries. The USB port can be used to continuously power the device, but it is not able to recharge the AA batteries. The USB-Power adapter is not required for the device and is not supplied. 3 AA batteries are supplied with the flowmeter. Question Q028 The Agilent G6691A Flowmeter OLED screen reads “Error code X.” What does that mean? Answer A028 All error codes are listed in the Operation Manual for the Flow Meter. Question Q029 The manual in softcopy states that the Agilent G6691A Flowmeter can communicate to our computer via a USB cable. Where is the driver to support the USB connection? Answer A029 See the ADM Flowmeter PC Connection Guide https://www.agilent.com/cs/library/usermanuals/public/Agilent%20ADM%20Flow%20Meter%20PC%20Connection%20User%20Guide.pdf Question Q030 There is a temperature sensitive sticker applied to the Agilent G6691A Flowmeter box. What is it for? Answer A030 The temperature sensitive sticker is intended for use during shipment. The Flowmeter is designed with the same operating and storage temperature limits as the GC. Question Q031 We activated the Agilent G6691A Flowmeter and inserted the new cartridge, but the display on the cartridge is showing a different time and date. Is this following USA timing? Answer A031 The Flow Meter is set on Coordinated Universal Time (UTC). The intent of the timer is to indicate one year of use from the installation dates rather than a specific start date. This satisfies the needs of many compliant departments worldwide. Question Q032 What are the part numbers for the Agilent G6691A Flowmeter? Answer A032 G6691A is the Flow Meter, which includes the first cartridge. G6692A is the replacement cartridge. Question Q033 What are the part numbers of the adaptors for the detectors for the Agilent G6691A Flowmeter? Answer A033 5190-9576 is the FID Adapter for Flow Meter 5190-9577 is the NPD Adapter for Flow Meter 5190-9578 is the TCD Adapter for Flow Meter Question Q034 What are the ranges of operating conditions for the Agilent G6691A Flowmeter? Answer A034 Flow Range: 0 to 750 ml/min Accuracy: 0-500ml/min +- 2% of reading or +- 0.2 ml/min whichever is greater, 501-750 ml/min +-3% of reading Operating temperature range: 0C to 45C Power: 3 AA batteries or USB power Question Q035 What do I do with the spare cartridge after it is out of calibration? Answer A035 The consumed cartridge can be disposed of per your local requirements for electronic waste. A new cartridge can be ordered online at https://www.agilent.com/en-us/products/lab-supplies/chromatography-spectroscopy/gas-management/adm-flow-meter#buy-products Question Q036 What gas line traps are recommended for the MSD's? Answer A036 https://www.agilent.com/en-us/support/mass-spectrometry/kb000405 Question Q037 What gas traps are recommended for Agilent GC? Answer A037 https://www.agilent.com/en-us/support/gas-chromatography/kb000299 Question Q038 What happens if I cover the vents on top of the Agilent G6691A Flowmeter? Answer A038 If the vents are covered on top, the Flow Meter will not read accurately and the internal components may be damaged, requiring the cartridge to be replaced. Question Q039 What happens if the Agilent G6691A Flowmeter overflows? Answer A039 If this does occur, the Flow Meter will log the date and time of when that occurred. A replacement cartridge will need to be ordered for the Flow Meter. Question Q040 What indicates the need To change the gas purifiers? Answer A040 https://www.agilent.com/en-us/support/gas-chromatography/kb001596 Question Q041 What is the maximum flow rate for the Agilent G6691A Flowmeter? Answer A041 The limit is 900uL per minute. If the customer hears a crunching noise, the transducer has been broken and the cartridge needs to be replaced. Question Q042 When I purchase the Agilent G6691A Flowmeter does it come with a flow cartridge? Answer A042 Yes. The replacement cartridge is not needed at time of purchase unless an extra cartridge is wanted as a spare. Question Q043 When we opened the box for the Agilent G6691A Flowmeter we noticed that the hose was not connected to the cartridge. The flowmeter is calibrated with the hose, so doesn’t it mean that the hose should already come pre-installed? Answer A043 The tubing used in calibration is identical to the tubing shipped separately in the box. A replacement part number is available if the tubing is damaged. The intent of this warning statement is to advise the customer against altering the tubing which will impact the calibration.
GC Columns
Question Q044 After replacing the column, what does error "Vacuum Pump is Not Ready" indicate? Answer A044 https://www.agilent.com/en-us/support/mass-spectrometry/kb001665 Question Q045 Can a Guard Column Be Utilized? Answer A045 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004806 Question Q046 Can graphite/polyimide ferrules be used with Pre-swaging tool Answer A046 https://www.agilent.com/en-us/support/gas-chromatography/kb005988 Question Q047 Can I order the old column cage design? Answer A047 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/can-i-order-the-old-column-cage-design Question Q048 Can I order these improved columns using the current part number? Answer A048 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/order-my-column-with-improved-performance-using-current-pn Question Q049 Can modules be connected To each other? Answer A049 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004808 Question Q050 Column Installation Pre-Swaging Tool - metal ferrules Answer A050 https://www.agilent.com/en-us/support/gas-chromatography/kb005989 Question Q051 Do I need a transfer line module for each column module? Answer A051 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004843 Question Q052 Do you have a GC Column Cross-Reference guide? Answer A052 https://www.agilent.com/en-us/support/gas-chromatography/gc-systems/7820a-gc/kb005556 Question Q053 Do you have information on GC Column test standards? Answer A053 https://www.agilent.com/en-us/support/gas-chromatography/kb005562 Question Q054 Does the new design affect the column performance? Can I use the same method? Answer A054 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/does-the-new-design-affect-the-column-performance Question Q055 GC Column Cross-Reference guide Answer A055 https://www.agilent.com/en-us/support/gas-chromatography/kb005554 Question Q056 GC Column Test standards Answer A056 https://www.agilent.com/en-us/support/gas-chromatography/kb005563 Question Q057 How do column modules connect to transfer line modules, the oven, the injector and detector? Answer A057 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004837 Question Q058 How do I select a Split/Splitless liner? Answer A058 https://www.agilent.com/en-us/support/gas-chromatography/kb005098 Question Q059 How many modules can be operate simultaneously with different temperature programs? Answer A059 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004807 Question Q060 How much force is needed to pre-swage a ferrule? Answer A060 https://www.agilent.com/en-us/support/gas-chromatography/kb005990 Question Q061 How should customer condition a PLOT column before use? Answer A061 https://www.agilent.com/cs/library/usermanuals/public/I-0300%20PlotColumns_029627.pdf Question Q062 I just received 2 columns that look very different. What should I do? Answer A062 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/i-just-received-2-columns-that-look-very-different Question Q063 Since this is a replacement, why are the part numbers different for the improved columns? Answer A063 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/different-part-numbers-for-the-improved-columns Question Q064 Spots on PLOT columns - will they affect the performance? Answer A064 https://www.agilent.com/en-us/support/gas-chromatography/gc-systems/7820a-gc/kb005539 Question Q065 Spots on PLOT columns and will they affect the performance? Answer A065 https://www.agilent.com/en-us/support/gas-chromatography/kb005540 Question Q066 What about the QC test? Will this be different? Answer A066 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/qc-test-will-this-be-different Question Q067 What are the possible causes for ghost peaks or carryover? Answer A067 https://www.agilent.com/en-us/support/gas-chromatography/kb001656 Question Q068 What are the possible causes for peak tailing? Answer A068 https://www.agilent.com/en-us/support/gas-chromatography/kb001654 Question Q069 What are the possible causes of no peaks being observed? Answer A069 https://www.agilent.com/en-us/support/gas-chromatography/kb000255 Question Q070 What are the Possible Causes of Peak Tailing? Answer A070 https://www.agilent.com/en-us/support/gas-chromatography/kb000254 Question Q071 What are the possible causes of peaks being larger and eluting earlier? Answer A071 https://www.agilent.com/en-us/support/gas-chromatography/kb000256 Question Q072 What are the possible causes of peaks of interest showing up in blank runs? Answer A072 https://www.agilent.com/en-us/support/gas-chromatography/kb000257 Question Q073 What are the possible causes of resolution loss and retention time shift after installing a new column? Answer A073 https://www.agilent.com/en-us/support/gas-chromatography/gc-systems/7820a-gc/kb005277 Question Q074 What are the Possible Causes of Resolution Loss and Retention Time Shift After Installing a New Column? - 5890 GC Answer A074 https://www.agilent.com/en-us/support/gas-chromatography/kb005279 Question Q075 What are the practical differences in the 3" (small) and 5" (standard) formats? Answer A075 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004834 Question Q076 What are the procedures for conditioning capillary GC columns? Answer A076 https://www.agilent.com/cs/library/datasheets/public/GCTroubleshooting_column_installation_notes.doc.pdf Question Q077 What are the steps to best troubleshoot column bleed? Answer A077 https://www.agilent.com/en-us/support/gas-chromatography/kb000258 Question Q078 What are the steps to condition a Capillary Column? Answer A078 https://www.agilent.com/en-us/support/gas-chromatography/kb002685 Question Q079 What are the steps to determine if a sample is causing damage to a capillary column? Answer A079 https://www.agilent.com/en-us/support/gas-chromatography/kb001439 Question Q080 What are the Steps to Install a Capillary Column in the PCOC Inlet? Answer A080 https://www.agilent.com/en-us/support/gas-chromatography/kb001988 Question Q081 What are the steps to install a capillary column into a TCD? Answer A081 https://www.agilent.com/en-us/support/gas-chromatography/kb002684 Question Q082 What are the steps to minimize and monitor column bleed? Answer A082 https://www.agilent.com/en-us/support/mass-spectrometry/kb002366 Question Q083 What are the steps to minimize problems with water injections on capillary columns? Answer A083 https://www.agilent.com/en-us/support/gas-chromatography/kb001433 Question Q084 What are the steps to obtain maximum performance when using a deans switch with PLOT columns? Answer A084 https://www.agilent.com/en-us/support/gas-chromatography/kb001872 Question Q085 What are the steps to reduce peak fronting? Answer A085 https://www.agilent.com/en-us/support/gas-chromatography/kb000253 Question Q086 What are the steps to troubleshoot a retention time shift? Answer A086 https://www.agilent.com/en-us/support/gas-chromatography/kb001655 Question Q087 What are the steps to troubleshoot baseline instability? Answer A087 https://www.agilent.com/en-us/support/gas-chromatography/kb001653 Question Q088 What causes peak fronting? Answer A088 https://www.agilent.com/en-us/support/mass-spectrometry/kb002650 Question Q089 What causes peak tailing? Answer A089 https://www.agilent.com/en-us/support/mass-spectrometry/kb002651 Question Q090 What causes the spots on PLOT columns and does it affect the performance of the columns? Answer A090 https://www.agilent.com/en-us/support/gas-chromatography/kb005537 Question Q091 What columns and flow rates are acceptable for proper operation of the MSD? Answer A091 https://www.agilent.com/en-us/support/mass-spectrometry/kb000396 Question Q092 What happens if I order using the current GC Column part number? Will I receive the improved columns? Answer A092 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/receive-column-with-improved-performance-using-current-pn Question Q093 What is "LTM" Technology? Answer A093 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004852 Question Q094 What is the correct distance a column should be installed in the GC Inlet? Answer A094 https://www.agilent.com/cs/library/quickreference/public/GC%20Column%20Installation%20Quick%20Reference.pdf Question Q095 What is the Maximum Operating Temperature for a Module? Answer A095 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004829 Question Q096 What is the procedure for installing a GC column in to GC-FID, GC-TCD, µECD and/or GC-MSD? Answer A096 https://www.agilent.com/cs/library/packageinsert/public/PN%20830-0120.pdf Question Q097 What types of GC Columns can be used? Answer A097 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004838 Question Q098 What would happen if I don't install the thermal shield? Answer A098 https://www.agilent.com/en-us/support/gas-chromatography/gc-gc-ms-technologies/ltm-series-ii-rapid-heating-cooling-for-7890-gc/kb004847 Question Q099 What’s the difference between the standard columns and the improved versions? Answer A099 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/difference-between-standard-columns-improved-versions Question Q100 Why are peaks of interest showing up in blanks? Answer A100 https://www.agilent.com/en-us/support/mass-spectrometry/kb000399 Question Q101 Why are the peaks smaller than expected (poor sensitivity)? Answer A101 https://www.agilent.com/en-us/support/gas-chromatography/kb005102 Question Q102 Why are there unwanted (ghost) peaks in the 7890A GC chromatogram? Answer A102 https://www.agilent.com/en-us/support/gas-chromatography/kb005100 Question Q103 Why do I occasionally get a PLOT column with what looks like a spot or spots in the column? What causes this? Will this affect the performance of the column? Answer A103 The spots you are seeing are small voids in the stationary phase coated inside of the tubing. It is normal to have some voids in the phase coating of PLOT columns. The columns are individually tested and verified to exhibit the correct chromatographic performance. Agilent guarantees the reliability and performance of our products with a 90-day warranty for all of our GC columns. Question Q104 Why does my column look different? Answer A104 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/why-does-my-column-look-different Question Q105 Why does the instrument have trouble tuning after a column replacement? Answer A105 https://www.agilent.com/en-us/support/mass-spectrometry/kb000401 Question Q106 Why improve HP-INNOWax and CP-Wax 52 CB columns? Answer A106 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/hp-innowax-cp-wax-52-cb-improvements Question Q107 Why is the effective internal diameter of the PLOT Column necessary when using the deans switch? Answer A107 https://www.agilent.com/en-us/support/gas-chromatography/kb001873 Question Q108 Will I experience any selectivity difference with the new GC Column? Answer A108 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/will-i-experience-any-selectivity-difference Question Q109 Will the EZ-Grip fit the updated column cage? Answer A109 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/will-the-ez-grip-fit-the-updated-column-cage Question Q110 Will the improved columns fit my regulated method? Answer A110 https://www.agilent.com/en-us/support/gas-chromatography/gc-columns/capillary/will-the-improved-columns-fit-my-regulated-method
General
Question Q111 What information is included on the CofA? Answer A111 A Certificate of Analysis is shipped with each standard detailing the specific MW characteristics for each vial. Question Q112 Where do I find the Certificate of Analysis for my product? Answer A112 https://www.agilent.com/search/?N=137&Ntt=cofa Question Q113 Where do I find the Material Safety Data Sheet (MSDS) for my product? Answer A113 http://www.chem.agilent.com/en-US/Search/Library/Pages/MsdsSearch.aspx
Inlets and Liners
Question Q114 What are the steps to chemically deactivate injection liners? Answer A114 https://www.agilent.com/en-us/support/gas-chromatography/kb002679 Question Q115 What are the steps to clean the GC injection port liner? Answer A115 https://www.agilent.com/en-us/support/gas-chromatography/kb000270 Question Q116 What indicates the liner or glass wool need replacement? Answer A116 https://www.agilent.com/en-us/support/gas-chromatography/kb000308 Question Q117 What indicates the need to change the septum or liner? Answer A117 https://www.agilent.com/en-us/support/gas-chromatography/kb000269 Question Q118 What is the appropriate amount of glass wool in a liner? Answer A118 https://www.agilent.com/en-us/support/gas-chromatography/kb000309
Lamps
Question Q119 Are there any performance compromises with multi-element lamps? Answer A119 https://www.agilent.com/en-us/support/icp-ms/kb005753 Question Q120 Are there storage guidelines for AA lamps? Answer A120 Hallow Cathode Lamps do not have an expiry date or shelf life. The lamp is totally sealed (like a light bulb) so even though it may be in storage for a while, we don’t expect to see any degradation in performance. You can check the lamp by checking the gain setting located on the lamp optimization screen in the software. If necessary check for an emission peak at the appropriate wavelength using a lamp emission scan. You should keep a record of the gain setting and confirm that it does not change significantly from the last test. A faulty lamp will show a glow from the cathode - but the gain setting will be very high - and the instrument will report an error message (either low lamp energy or it may show a "peak not found" error). There is also an AA at Work publication (application note) that may be a useful reference as it provides more detail to help users learn more about the design characteristics and operation of using hollow cathode lamps and deuterium lamps. https://www.chem.agilent.com/Library/applications/aa083.pdf Question Q121 Can all lines be used in multi element lamps without interference? Answer A121 https://www.agilent.com/en-us/support/icp-ms/kb005754 Question Q122 Can spectral interferences be overcome? Answer A122 https://www.agilent.com/en-us/support/icp-ms/kb005756 Question Q123 Do any lamps have solid cathodes? Answer A123 https://www.agilent.com/en-us/support/icp-ms/kb005751 Question Q124 Do other elements in the cathode affect the lifetime? Answer A124 https://www.agilent.com/en-us/support/icp-ms/kb005752 Question Q125 Does the flickering glow around the anode mean that the lamp is faulty? Answer A125 https://www.agilent.com/en-us/support/icp-ms/kb005747 Question Q126 How do I dispose of my hallow cathode lamp? Answer A126 You can dispose of old or used hollow cathode lamps in accordance with local EPA or Government regulations for regular vacuum lamps (or globes). Hollow cathode lamps contain small quantities of chemicals (usually less than 5 grams per lamp) and the lamps should be disposed of in a responsible manner because of these chemicals. Question Q127 How do you know when a lamp is worn out? Answer A127 https://www.agilent.com/en-us/support/icp-ms/kb005759 Question Q128 How much fill gas is added? Answer A128 https://www.agilent.com/en-us/support/icp-ms/kb005749 Question Q129 Is there a compromise in intensity with the Ag/Cd/Pb/Zn or Al/Ca/Mg multi elements lamps? Answer A129 https://www.agilent.com/en-us/support/icp-ms/kb005757 Question Q130 What additional information is available on lamp construction and operation? Answer A130 https://www.agilent.com/en-us/support/icp-ms/kb005761 Question Q131 What are the steps to perform a DAD intensity test using the ChemStation? Answer A131 https://www.agilent.com/en-us/support/liquid-chromatography/kb002436 Question Q132 What are the steps to perform a DAD wavelength calibration test using a control module? Answer A132 https://www.agilent.com/en-us/support/liquid-chromatography/kb002444 Question Q133 What are the steps to perform a DAD wavelength calibration test using the ChemStation? Answer A133 https://www.agilent.com/en-us/support/liquid-chromatography/kb002443 Question Q134 What are the Steps to Perform a FLD Dark Current Test using the ChemStation? - 1100 Series LC Answer A134 https://www.agilent.com/en-us/support/liquid-chromatography/kb002445 Question Q135 What are the steps to perform a FLD lamp intensity test using the ChemStation? Answer A135 https://www.agilent.com/en-us/support/liquid-chromatography/kb002446 Question Q136 What are the steps to reset the lamp hours on a DAD, MWD or VWD? Answer A136 https://www.agilent.com/en-us/support/liquid-chromatography/kb000043 Question Q137 What are the typical % Gain or EHT values for hollow cathode lamps? Answer A137 https://www.agilent.com/en-us/support/icp-ms/kb005762 Question Q138 What are the typical life times of a cathode lamp? Answer A138 https://www.agilent.com/en-us/support/icp-ms/kb005760 Question Q139 What are typical results from a Arsenic (As) hollow cathode lamp drift test? Answer A139 https://www.agilent.com/en-us/support/icp-ms/kb005763 Question Q140 What causes baseline noise and drift after replacing the DAD lamp and flow cell? Answer A140 https://www.agilent.com/en-us/support/liquid-chromatography/kb001396 Question Q141 What indicates the FLD lamp is on? Answer A141 https://www.agilent.com/en-us/support/liquid-chromatography/kb000051 Question Q142 What is the approximate life of UV Detector lamps? Answer A142 https://www.agilent.com/en-us/support/uv-vis-uv-vis-nir/kb002479 Question Q143 What is the approximate lifetime of UV detector lamps? Answer A143 https://www.agilent.com/en-us/support/liquid-chromatography/kb002211 Question Q144 What is the black spot? Answer A144 https://www.agilent.com/en-us/support/icp-ms/kb005745 Question Q145 What is the purpose of the fill gas? Answer A145 https://www.agilent.com/en-us/support/icp-ms/kb005748 Question Q146 What type of testing do lamps undergo during production? Answer A146 https://www.agilent.com/en-us/support/icp-ms/kb005750 Question Q147 Why do lamps wear out? Answer A147 https://www.agilent.com/en-us/support/icp-ms/kb005758 Question Q148 Why does a new lamp look used? Answer A148 https://www.agilent.com/en-us/support/icp-ms/kb005744 Question Q149 Why does my hallow cathode lamp looked burned inside? Answer A149 In the lamp purification and processing operation, the polarities of the cathode and the anode in the lamp are reversed so that the zirconium anode now becomes the cathode. This subjects the zirconium anode to ion bombardment. During this discharge a small amount of the zirconium anode material is vaporized and deposited on the inside envelope of the lamp. This is the dark film visible on the glass envelope near the anode. Zirconium is a very active “getter” for impurity gases such as oxygen and hydrogen, and this discharge serves to purge the lamp of these impurity gases. That is, this activated film of zirconium will absorb any impurity gases that may have escaped the previous purification stage. This “black patch” that is visible on the inside of the lamp ensures that the fill gas is free of impurities and contributes to the long shelf life of Agilent lamps – and also ensures continued spectral purity throughout the life of the lamp. Finally the lamp is then filled with spectroscopically pure gas and sealed. Processed lamps are then operated for several hours prior to testing. The extended processing cycle and use of specially pure materials ensures dependable performance from Agilent hollow cathode lamps. For further details on the features and characteristics of the hollow cathode lamp, please refer to the Agilent Application Note titled “Features and Operation of Hollow Cathode Lamps and Deuterium Lamps” available from the Agilent literature library on the Agilent website www.agilent.com/cs/library/applications/aa083.pdf Question Q150 Why treat the cathode? Answer A150 https://www.agilent.com/en-us/support/icp-ms/kb005746
LC Columns
Question Q151 Agilent AdvanceBio Peptide Mapping Columns Specifications Answer A151 https://www.agilent.com/en-us/support/liquid-chromatography/kb005975 Question Q152 Agilent Bio IEX Column Specifications Answer A152 https://www.agilent.com/cs/library/datasheets/public/5991-6120EN.pdf Question Q153 Agilent Bio SEC-3 Column Specifications and Options Answer A153 https://www.agilent.com/cs/library/selectionguide/public/5990-9384EN.pdf Question Q154 Agilent Bio-Monolith Column Specifications Answer A154 https://www.agilent.com/cs/library/datasheets/public/5991-6040EN.pdf Question Q155 Agilent ZORBAX Bio SEC-5 Column Specifications and Options Answer A155 https://www.agilent.com/cs/library/selectionguide/public/5990-9384EN.pdf Question Q156 Bio LC Column User Guides Answer A156 https://www.agilent.com/en-us/support/liquid-chromatography/kb005960 Question Q157 Can I really effectively use these very short columns on my HPLC instrument? Answer A157 Yes, you can. For columns of 3.0 mm i.d. and above, no instrument adjustments are necessary. For gradient separations with 2.1 mm and 1.0 mm i.d. columns the ideal HPLC is a high-pressure mixing instrument — like the Agilent 1100 HPLC with the binary pump, because it minimizes the gradient delay volume. Using a low volume mixer and the injector by-pass (or micro-injector) further minimize gradient delay volume. Narrow i.d. tubing and a low volume detector flow cell are preferred but not necessary. These changes are easy to make and allow you to effectively use columns as small as the 2.1 x 30 mm columns or even the 2.1 x 15 mm columns Question Q158 Do you think that using toluene as a standard for measuring GPC column efficiency is appropriate? Answer A158 Yes. Conditions for this evaluation are chosen so that the small molecule toluene does not interact with the packing, but diffuses through all the pores of the GPC column packing. By conducting the experiment in this way, the width of the eluting peak is a function of the particle size of the packing and how well the column bed is packed. That's what you want to know when you are measuring efficiency. This sample is commonly used and recommended by most manufacturers to test efficiency of these columns. Alternate standards may be used to determine other characteristic of a GPC column, e.g., exclusion volume. Question Q159 How can one properly wash GPC columns (safely)? Answer A159 Wash the column in the reverse direction, not attached to the detector and at half the recommended flow rate (keep the pressure below the recommended maximum). First choose a solvent that will dissolve what you believe has contaminated the column. Most GPC columns are PS-DVB and you need to check the solvent compatibility before using a solvent. Many wash solvents are a higher viscosity than the typical eluting solvents so a lower flow rate with careful attention to pressure is needed. Anionic samples can adsorb onto PS-DVB and if these have contaminated your GPC column a wash solution with a salt is recommended. Check to see what types of salts are recommended for the column. In some cases the polarity of the material adsorbed may require washing with organic solvents modified with acid (formic or acetic) or base (triethanolamine) (check the pH range) or some water may be compatible with an appropriate organic solvent. If more hydrophobic material were retained, elevated temperature along with an appropriate organic solvent would be recommended. Once again you need to check the maximum temperature range allowed for your column. If you wash carefully the column should not degrade from the solvent switching. Question Q160 How do I calculate conditions for my Prep LC column? Answer A160 https://community.agilent.com/docs/DOC-1951/edit?ID=1951&draftID=4284# Question Q161 How do I perform column conditioning for IP391/07 Diesel analysis? Answer A161 The column sequence should be SB-CN first and Zorbax NH2 second. The NH2 column needs a conditioning step before use according the attachment. The used solvent must be extremely dry to avoid retention time and separation changes. Question Q162 How do you clean C4 column? Answer A162 Much as you would any RP-phase column and I've included some general instructions provided that you are using standard mobile phase conditions to separate small molecules. If you are separating peptides and proteins or if your sample is dissolved in plasma, then the guidance is different and I recommend that you call 800-227-9770, press option 1 and ask for HPLC column technical support for more information (within the US, other areas contact your local Agilent sales office. Question Q163 How do you evaluate for column voids? Answer A163 First check your method and find out if you have operated at a pH higher than is recommended for the column. This is probably the major cause of column voids because silica can dissolve causing the column void. The sample injection solvent as well as the mobile phase needs to be considered here. If there is a column void, you will see a change in peak shape - tailing, broadening, or split peaks - on every peak in the chromatogram. A void will not usually cause a change in only one peak in the chromatogram. It also does not typically cause a change in analyte retention. You can also turn the column around and if there is a column void then the peak shape should be poor in the reverse direction as well. The only definitive way to check for a column void is to open the column and this should be done only as a last resort in identifying the problem with the column. Question Q164 I can’t get enough retention of my basic compounds at low pH, what column should I try next? Answer A164 The Eclipse XDB column can be used in the intermediate pH region — from pH 3 - 8 for the longest lifetime. The primary advantage of using the intermediate pH region is a possible increase in retention for basic compounds. Most basic compounds have pKa values of 5 or greater, therefore at pH 6 - 8 some of these compounds may become non-charged and the column will retain them more. In addition, at pH > 5 the residual silanols on the silica surface will become charged. This can lead to stronger interactions with basic compounds and increase retention. The Eclipse XDB column is the ideal column to use in this mid-pH region because the dense-bonding and double end capping will cover the most active silanols on the surface of the column, and any residual silanols can contribute to increased retention of basic analytes without causing excessive peak tailing. Figure 1 shows a plot of retention vs pH for a group of basic compounds. At pH 6 and higher the retention of all of these compounds increases due to increased interaction with the column, though at pH 6.5 only Triprolidine is non-charged. If retention has not increased enough, then the next step would be to try the Extend-C18. All of the basic compounds shown in Figure 1, except Triprolidine, have pKa values of 9.0 - 9.2. Therefore they must be analyzed at pH 10 or higher before they are non-charged. The Extend-C18 can effectively be used at this high pH to improve the retention of basic compounds. Question Q165 I typically select 4.6 x 250 mm, 5 µm columns for my analytical work because I have complex samples, but I need to reduce my analysis time and increase my sample throughput. What can I do? Answer A165 Rapid Resolution columns are ideal for your needs. A Rapid Resolution 4.6 x 150 mm, 3.5 µm column will reduce your analysis time by 40% while maintaining your resolution. Your gains are in reduced analysis time, whether you are doing isocratic or gradient separations, and substantial solvent savings (Figure 5). You may be able to choose even shorter Rapid Resolution column lengths and maintain the desired resolution (Rs) because Rs a N1/2. This means that decreasing column length, and therefore efficiency, will not decrease Rs by the same amount. It is likely that resolution will be maintained on even shorter column lengths and choosing 75 mm or even 50 mm column lengths can reduce analysis time even more. Question Q166 I’m currently using a C18 column and my separation has a couple of peaks that elute early and a couple of peaks that elute late. What can I do to reduce the analysis time and maintain resolution of the early eluting peaks? Answer A166 First, you could try a gradient elution method on the C18 column. But many people do not like to use gradients, so choosing a different bonded-phase may help. Short chain polar bonded-phases such as the SB-CN and SB-Phenyl are ideal for separations like this. The increased polarity of these phases reduces retention of the later eluting hydrophobic compounds while often maintaining the retention of earlier eluting hydrophilic compounds. Figure 2 shows this clearly. Using the same mobile phase conditions these three compounds are well resolved on the SB-CN in about 5 minutes. The same separation on the SB-C8 column takes nearly twice as long and provides incomplete resolution. Question Q167 I’m trying to separate some very difficult basic compounds at pH 2.5 using a C18 column and their tailing factors are still too high — around 2.0. Which column should I try next? Answer A167 The Bonus-RP column would be a good column to try. This column has an amide group in the alkyl chain. This amide group reduces interactions between basic compounds and the residual silanols on the silica surface by acting as an internal competing base and reducing peak tailing of basic compounds. The Bonus-RP column, while ideal for the mid-pH range where silanol interactions are more likely, will also improve peak shape at low pH. The Bonus-RP column uses the same sterically protecting bonding used for StableBond columns. This improves the lifetime of the Bonus-RP column at low pH and makes it a good choice in situations like this. Question Q168 If one peak in a chromatogram is tailing but the others are not, what is the likely cause? Answer A168 Peak tailing can be caused by a variety of reasons and I would prefer to ask you several questions about your sample before I submit a response. Since most of the peaks in your chromatogram are well shaped and only one is tailing, I suspect that the chemistry of the column and sample are such that a secondary interaction is inducing the tailing. Modifying the mobile phase or selecting another HPLC column can reduce these secondary interactions. Question Q169 I'm having a problem with resolution and selectivity for a phenyl column using different lots, how can I resolve this problem? Answer A169 First, evaluate the methods used on each column. Are their any differences in the column histories or column equilibration procedures? Differences in either of these may alter the behavior of analytes on a column. Recently someone had a problem with a change in resolution and selectivity on two phenyl columns, which was resolved by flushing the column once again with 100% methanol. This may make a good starting point for you. In the case I just described we had a third column lot to compare it to so that when we flushed the column and retested the sample in its mobile phase we achieved the expected results and knew that this flush should be added to the equilibration procedures. Our typical recommendation would be to flush the column with 100% acetonitrile or methanol, which ever is used as the organic modifier in the mobile phase. In the case I described, only the methanol changed the column, possibly due to interactions with the phenyl rings in the bonded-phase. You should also make up fresh mobile phase and test it in case the problem occurred with the mobile phase and make sure the instrument is operating as expected. These steps will eliminate some of the potential sources of column-to-column variations. The next step is to review your method ruggedness. Are any of your analytes sensitive to pH or ionic strength? If slight changes in pH or ionic strength affect retention then you may need to modify your method and select conditions where changes in these parameters do not affect resolution and selectivity as much. Acidic and basic analytes can be very sensitive to pH and ionic strength changes around the pKa of the analyte. Finally ask the column manufacturer for assistance. They may be able to supply additional columns from older lots that have worked or from newer lots while the method is evaluated. The manufacturer may also be able to make some suggestions based on your method parameters. Question Q170 I'm working with environmental samples that may contain sulfur. I suspect that a few samples insufficiently cleaned up. Column performance has declined significantly. Is there any way to restore performance? Answer A170 I assume you are using a reversed-phase column. Please review the recommended HPLC column cleaning procedure . I do suggest that if column performance declined quickly that you consider using a guard column and possibly consider including a preliminary clean-up step before injection. Question Q171 Is a Rapid Resolution 3.5 µm column more likely to fail than 5 µm column? Answer A171 No. The Rapid Resolution columns are as rugged as the 5 µm, 250 mm columns. There are two types of accelerated column failure attributed to using smaller particles. First, smaller particle columns are thought to plug faster. This is not true when a standard 2 µm frit is used at the top of the column. Because of careful particle size control and the use of 3.5 µm particles, ZORBAX Rapid Resolution columns will not contain any particles as small as 2 µm. This means a standard frit can be used on the column and the Rapid Resolution columns will be no more prone to plug than a 5 µm particle size column. Second, columns with smaller particles are thought to have shorter lifetimes because the column beds compress, leaving voids that cause peak broadening and tailing. It is true that 3.5 µm particle size columns will operate at slightly higher pressures than 5 µm columns, but ZORBAX particles can easily withstand these increases in pressure. ZORBAX particles are packed at 8000 psi and can easily withstand pressures up to 5000 psi in routine use. A Rapid Resolution 4.6 x 150, 3.5 µm column will typically be operated below 3000 psi, so the column bed will not compress when using ZORBAX Rapid Resolution columns. So both the rugged ZORBAX particles and the standard 2 µm column frit assure you long column lifetime when using Rapid Resolution columns. Question Q172 Is there a way to separate Copper galacturonate from Copper gluconate?. Answer A172 I suggest trying ZORBAX SAX 70Å 5um, 4.6 x 150mm 820888-901, with water and acetonitrile mobile phase, say 50% to start. Hopefully the two -OH rich compounds will be retained on the quaternary amines of the SAX column, adjust water strength to elute. UV detection (230 nm) should work since they both have a carbonyl. Question Q173 Is there any kind of "generic" column test to track the performance of a column? Answer A173 The best way to evaluate a columns performance is to use the QC test that should be shipped with each HPLC column provided to you. By comparing the efficiency, retention and peak shape of the peaks in the sample and very importantly the pressure under these experimental conditions, you will be able to tell if your column has changed over time. These tests can also tell you how your column has changed, as these are the same tools we use to diagnose a column problem. Significant changes in retention, can suggest loss in bonded phase and significant changes in peak width/efficiency and pressure can suggest column contamination. Dramatic losses in peak width and efficiency suggest a column void. Most manufacturers use similar solutes and mobile phase test conditions to QC their columns. However, different bonded phase require modification of the amount of organic modifier to achieve reasonable retention. What's important is that that you QC-test a column on your system before you use it for a particular project. Then you know how that column performs on your system, before you start injecting samples and before you start having problems. If problems develop, use this QC-test to verify that the HPLC system and the column are performing well. Before you retire a column from a project, clean and QC-test the column. If for any reason, you start a project with a used column, at a minimum, QC-test the column before injecting samples. Question Q174 LC & LC/MS Columns - USP Designations Answer A174 https://www.agilent.com/en-us/support/liquid-chromatography/kb005979 Question Q175 LC Column User Guides Answer A175 https://www.agilent.com/en-us/support/liquid-chromatography/kb005965 Question Q176 My work requires very high sample throughput. How fast can I do a gradient separation? Answer A176 More and more people, especially those who analyze combinatorial chemistry samples, need to analyze a lot of unknown samples quickly. The best way to do this is on short, Rapid Resolution columns with rapid gradient times. The best gradient time is the one that resolves all of your analytes in the least amount of time. On very short columns — 30 and 50 mm lengths — a good starting point would be a 2 - 5 minute gradient. From there optimize your separation for organic range and gradient time. On these short columns you can easily increase the flow rate to further decrease analysis time to 30 seconds without exceeding the pressure limits of the columns. Column re-equilibration times are typically as short as the analysis times — 3 - 5 minutes for the 50 mm columns. Question Q177 Should I just select the Bonus-RP column for all my method development with basic compounds? Answer A177 Basic compounds are best analyzed following the method development scheme outlined earlier in this brochure. Select an SB-C8 or SB-C18 column for initial development and use a buffered low pH mobile phase. Many times this approach provides a good separation and the StableBond columns will have exceptional lifetime at low pH even at high temperatures. Question Q178 Some manufacturers’ indicate (newer) columns can be reversed to remove blockage from the front of the column. Is this generally recommended or does it depend on the column? Answer A178 HPLC columns are more efficient and packed better than years ago, therefore this is generally recommended for most reversed-phase and normal phase silica based columns. This procedure is designed to remove particles from the column frit when high pressure occurs at the column inlet, but it will not work all the time. Because it does not require opening the column it is worth trying. Make sure when you do this that you disconnect the column from the detector and make sure the particles that plugged the column are not coming from the HPLC system or you may just plug the frit at the back end of the column and this may not be replaceable. Columns can also be turned in the reverse direction for washing/cleaning with stronger solvents to remove adsorbed material. This has the benefit of not exposing the rest of the column to the contaminants. When this is done the column should also not be attached to the detector. Question Q179 There are a lot of different column configurations (dimensions) available, but I don’t see the one I’m looking for. Can I get a column made in the configuration I want? Answer A179 Most likely. Agilent is continually adding to its HPLC column offerings, so check with your Agilent column distributor or in the US call Agilent Customer Assist at 800-227-9700 and ask for HPLC column support to find out if the column configuration you need is currently available. For locations outside the U.S., your authorized Agilent column distributor, listed on the Agilent website, can also help you find out the cost and delivery time. If the column is not available, a column can be packed in the configuration of your choice with available bonded-phases. Select an internal diameter and specify a column length, bonded-phase, and particle size using product number 899999-999. A special products quote (SPQ) will then be issued to track your order. Question Q180 USP designated LC Columns - 1050 LC Answer A180 https://www.agilent.com/en-us/support/liquid-chromatography/kb005572 Question Q181 Very short narrow-bore columns seem ideal for LC/MS. What bonded phase should I use? Answer A181 LC/MS requires volatile mobile phases and the most suitable buffers are acetate and format. The buffer range for acetate is from pH 3.8 - 5.8 so the best bonded-phases to start with would be the Eclipse XDB C18 and C8. Formate is also widely used and has a buffer range from pH 2.8 - 4.8. When using this buffer you can select StableBond columns then Eclipse XDB columns and follow the same method development scheme as discussed in the method development section. For high pH, using ammonium hydroxide at pH 10.5, select Extend-C18. Question Q182 What are the steps to troubleshoot leak errors? Answer A182 https://www.agilent.com/en-us/support/liquid-chromatography/kb002787 Question Q183 What column do you recommend to analyze raw materials? Answer A183 Column choice, and appropriate mobile phase, depends on many characteristics of the sample such as polarity, pKa of ionizable functional groups, solubility vs. pH, and molecular weight. Question Q184 What do you recommend doing when a impurity peak starts to co-elute and actually starts to make itself present in your main peak as a hump to the peak? Answer A184 How much of a change in resolution are you seeing? Has resolution decreased over time or has the separation been disappointing right out of the box? Do you have any information telling you what factors may cause these peaks to converge? Are they sensitive to small changes in temperature, pH or organic composition? Is one compound much larger than the other? These factors and more can be causing the two peaks to merge. More information would be required to specifically answer this question. Question Q185 What do you think about neutralizing extra-column volume by packing short columns with wide-bores (7mm)? Answer A185 As long as you are willing to accept operating at a higher flow rate. If you operate at 1.0 mL/min on a 4.6 mm i.d. column, then operating at 2.3 mL/min would be the equivalent linear velocity when using a 7.0 mm i.d. column. Solvent waste may be an issue. Question Q186 What does the filter do that the guard-column does not? Answer A186 Nothing really - both a filter and a guard can capture small particulates that are either coming from your sample, mobile phase or are from wear-and-tear of instrument seals and gaskets. A guard, on the other hand, traps sample material that is strongly retained on your column, material that would not elute from the column under the experimental conditions defined by the method. These strongly retained materials can build over time at the head of the column and generally lead to high pressure and/or poorly shaped peaks. In some cases, these retained materials cannot be removed even after extensive washings with strong solvent. Question Q187 What does the Flexible Stainless Steel Capillary Tubing Color-Coding Indicate? - 1090 Series LC Answer A187 https://www.agilent.com/en/support/liquid-chromatography/kb000052 Question Q188 What is meant by peak plate? Answer A188 Plates" allow you to compare the efficiency of different columns and its measure is determined by the width of a peak in a chromatogram relative to its retention. In general, for the same retention , the more narrow a peak, the more efficient the column and the more efficient the column has more "plates". HPLC instruments often determine column plates or efficiency by measuring the peak width , typically at half-height, using the following equation:N = efficiency = plates = 5.54(tR/w1/2)2 ,w1/2 = peak width at half-height in min.,tR = retention time in min. The efficiency of a column is reported by the manufacturer of your column and is generally provided in a column report. For a more detailed discussion on how plates and column efficiency relate to resolution and your analyses refer to the text Practical HPLC Method Development written by Lloyd R. Snyder, Joseph J. Kirkland and Joseph L. Glajch, Wiley-Interscience , Second Edition, 1997. Question Q189 What is the purpose of the H3PO4 wash of the column? Answer A189 A phosphoric acid wash has been shown to be effective at reducing tailing caused by the sample complexing with metals in the HPLC system. Typically a 1% phosphoric acid wash of the system and column is suggested to eliminate this tailing and it works. It is perfectly reasonable to use these recommended wash conditions with Agilent ZORBAX StableBond reversed-phase HPLC products. The StableBond HPLC column is particularly stable at low pH - the SB-C18 column is stable at a pH of 0.8 and 90°C. How can you tell if your peak tailing is caused by metal complexation? Look to see if a lone pair of electrons on either a N or O atom can chelate with the metal to form a 5 or 6-membered ring. Metal complexation is a commonly overlooked cause of peak tailing and metals are presents in every HPLC system. Question Q190 What might cause a rhythmic, wavy baseline? Answer A190 This is rarely a column problem. The most common causes of this problem are related to the pumping system. a) If you have an isocratic pump, your pump seals may be worn. Typically your pump will have two pistons and seals. One may be more worn than the other causing flow and pressure variations in a very rhythmic pattern, which will also be seen by the detector. You can confirm a pump seal problem by varying the flow rate. The frequency (in time) of the baseline fluctuations will increase/decrease proportionally to an increase/decrease in the flow rate if the pump seals are worn. The solution is to replace the pump seals. b) If you have a multiple pump system (binary, ternary, or quaternary), the problem may also be pump seals but could also be a function of insufficient mixing or of a malfunctioning proportioning valve. The former can be solved by adding a mixing column (ask your manufacturer for a recommendation) and the latter by a service call. Another possibility is an aging lamp although these signal fluctuations are not as perfectly periodic as a pump problem. An aging lamp can be distinguished from a pump problem with the flow rate experiment described above. If the frequency of signal fluctuations (in time) don't change when the flow rate is varied, the problem may be the lamp. Question Q191 What solvents do you recommend to clean the columns? Answer A191 I generally recommend using the organic solvent acetonitrile to clean reversed-phase, C8 or C18, silica-based HPLC columns. THF is a strong reversed-phase organic solvent and that may be difficult to remove from hydrophobic bonded-phases, causing chromatographic variability. For most situations, acetonitrile is a strong enough solvent to remove sample and mobile phase components that have accumulated on your column during routine isocratic use. Before cleaning the column with organic solvent however, be sure that you have removed any buffer salts that may be in your mobile phase and column to avoid precipitation of these salts in the column. To do that, prepare your mobile phase without buffer salts and flush your column with 15-20 column volumes of this mixture. That is, if your mobile phase is fifty-percent acetonitrile and fifty-percent 50 mM phosphate buffer, then the appropriate first step would be to remove the buffer salts with 50/50 mixture of acetonitrile and water. Then I would suggest flushing the column with pure acetonitrile thoroughly for 30-40 column volumes. If you believe there may particulates at the top of the column, e.g., pressure may be higher than normal, you should consider disconnecting the column from the detector and reversing the direction of flow before the cleaning step. Do no reconnect the column to the detector, but allow the cleaning solvent, in this case, acetonitrile, to collect in a beaker. Since acetonitrile is commonly recommended for long-term column storage, the column can be easily stored after cleaning using the procedure described above. When ready to re-use this stored solvent, remember to flush the column with your mobile phase composition without buffer salts before introducing the buffered mobile phase, again to reduce the possibility of salt precipitation. Question Q192 What temperature should I use for my separation? Answer A192 Temperature control for separations is important for long-term retention reproducibility, one factor of method ruggedness. Controlling temperature at 35 - 40°C is normally sufficient for good method reproducibility and ruggedness. In addition, the use of elevated temperature can have other benefits. First, it reduces the system operating pressure by reducing the viscosity of the mobile phase. Second, it will reduce analysis time, which can substantially increase productivity. Third, temperature may change the selectivity of a separation. Not all compounds have the same response to temperature so the selectivity of a separation can change dramatically when temperature is increased or decreased. StableBond columns have high temperature limits — SB-C18 can be taken up to 90°C, at low pH, and SB-C8 , SB-Phenyl , SB-CN and SB-C3 can be taken up to 80°C, making it possible to optimize your separation without changing columns, particularly if you are analyzing ionizable compounds. Question Q193 What types of column are dissolved above pH 7? Answer A193 Many commercially available silica-based HPLC packings dissolve at pH 7, for the simple reason that silica is soluble at pH 7. Therefore, Agilent has developed technology to significantly reduce silica dissolution and extend the usable pH range of silica-based columns. The ZORBAX Extend-C18 can be effectively used above pH 8 and even up to pH 11.5. Bidentate-bonding is key to this enhanced stability while offering the high efficiency that only silica can provide. Question Q194 What's the best procedure for cleaning out ion pair reagent materials from a column? Answer A194 Some detergents are used as ion-pair (IP) reagents, e.g. SDS with a carbon chain length of twelve. However, most commonly used ion-pair reagents have shorter carbon chains, e.g. hexane sulfonate has carbon chain length of six. IP reagents with longer carbon chains are more difficult to remove from RP-HPLC columns. Studies have shown that ion-pair reagents and some detergents are best removed with long washes with a 50/50: v/v, methanol/water mobile phase system. If you can regain the retention, selectivity and efficiency (resolution) of a known separation (e.g., QC sample) after washing, then you might have successfully removed the ion-pair reagent. I would not recommend using this column for developing a new method as exposure to ion-pair reagents may change the retention characteristics of a column permanently. If scouting runs for a new method are conducted on a column exposed to ion-pair reagents, a new column should be purchased as soon as possible to verify that the selectivity is not different. In my opinion, columns exposed to IP reagents should be dedicated to the ion-pair method. Question Q195 When do you see peak fronting? Answer A195 Peak fronting can occur under a variety of conditions. The most well known is peak fronting that occurs due to column overload. In this case you will typically see the peak retention time shift slightly shorter. This is also the simplest to evaluate, because you can just inject less sample and evaluate the peak shape. But there are at least four other causes of peak fronting. These are column channeling, ionic interactions between the analytes and the silica, poor solubility of the analyte in the mobile phase, and wettability problems between the mobile phase and bonded-phase. A problem with column channeling should result in peak fronting on all peaks or at least the largest peak in the chromatogram if the others are very small. If this occurs you need to replace the column. To evaluate this problem try a new column. Increasing the buffer ionic strength or changing the pH of the mobile phase can often improve ionic interactions, which cause fronting. Increasing the ionic strength can reduce interactions between the silica and ionic analytes and changing the pH can have the same effect. Solubility needs to be assessed by trying to improve the solubility of the analyte and evaluating the resulting chromatogram. For instance, you can increase the time you sonicate your sample and re-inject or dissolve it in a solvent where the sample has good solubility and then dilute and inject in the mobile phase. Also any sample you suspect has a solubility problem should be filtered. You can also mix your sample and mobile phase off-line to see if it is visibly soluble in the mobile phase. Wettability refers to the ability of the mobile phase to fully penetrate the bonded-phase so that analytes interact with all of the bonded-phase. In cases where there is a highly aqueous mobile phase with a C18 column complete wettability may not be achieved. The bonded-phase can fold over on itself. The results may be loss of retention and distortions in peak shape. If you can, increase the amount of organic in the mobile phase and re-evaluate peak shape. If not you may need to consider selecting a column designed for use in very high aqueous mobile phases. These are the typical causes of peak fronting and some solutions Question Q196 When should I select Phenyl bonded-phases? Answer A196 Both SB-Phenyl and Eclipse XDB-Phenyl bonded phases provide unique selectivity and are often a good choice for changing the selectivity for two closely eluting analytes. Phenyl bonded phases are less retentive than C8 bonded-phases so they offer a second option for reducing retention of late eluting hydrophobic compounds and minimizing analysis time Question Q197 Where can I find the LC Column User Guides? Answer A197 http://www.agilent.com/en-us/support/liquid-chromatography/kb005965 Question Q198 Why are there ghost peak In my chromatogram? Answer A198 https://www.agilent.com/en-us/support/liquid-chromatography/kb000047
LC Pumps
Question Q199 At what interval should the pump seals be replace? Answer A199 https://www.agilent.com/en-us/support/liquid-chromatography/kb000067 Question Q200 At what interval should the pump seals be replaced? Answer A200 https://www.agilent.com/en-us/support/liquid-chromatography/kb000025 Question Q201 What are the steps to replace the pump seals? Answer A201 https://www.agilent.com/en-us/support/liquid-chromatography/kb000057 Question Q202 What is the break / wear - in procedure for piston seals? Answer A202 https://www.agilent.com/en-us/support/liquid-chromatography/kb000050 Question Q203 What lubricant should be used on the base of Sapphire Plungers? Answer A203 https://www.agilent.com/en-us/support/liquid-chromatography/kb005126 Question Q204 What types of injector rotor seals are available? Answer A204 https://www.agilent.com/en-us/support/liquid-chromatography/kb005116
Sample Preparation
Question Q205 Are any additional accessories or parts included with the purchase of a PPM-48 (p/n 5191-4101)? Answer A205 Yes. The PPM-48 is shipped with: • Waste rack with three waste bins p/n 5191-4112 • Installation kit p/n 5191-4114 • User Manual publication number 5991-8133EN The manifold comes with a sealing gasket already installed. Question Q206 Are any additional accessories or parts included with the purchase of a PPM-96 (p/n 5191-4116)? Answer A206 Yes. The PPM-96 is shipped with: • Single well waste plate p/n 5191-4121 • Holder for plates p/n 5191-4120 • Installation kit p/n 5191-4114 • User Manual publication number 5991-8133EN The manifold comes with a sealing gasket already installed Question Q207 Are there any consumables or parts that I need to regularly order or replace? Answer A207 The only part that needs to be regularly replaced (both units) is the sealing gasket (PPM-48: 5191-4110; PPM-96: 5191-4117). Please note that each processor has its own unique sealing gasket (different p/n- because they are different shapes). Agilent recommends ordering these parts as needed due to adhesive expiration dates. The PPM-48 also uses disposable waste bins (p/n 5191-4113) to collect waste during various sample processing steps. They can be reused (customer's best judgment should be used to determine when they need replacement- depends on solvent(s) used, hazard of the sample/solvent, wear and tear, etc.). The single well waste plate for the PPM-96 (p/n 5191-4121), like the PPM-48 waste bins, is also a consumable, and while it can be reused, it should be replaced as necessary. Question Q208 Does the processor require any maintenance and if so, how often should it be performed? Answer A208 The sealing gasket should be wiped periodically with methanol and a lint free cloth. For simplicity, this can be done after every extraction procedure is completed. The sealing gasket should also be replaced 1-2x per year depending on wear, solvent usage, and overall condition. Instructions for replacement are included with each new sealing gasket and are also found in the user manual (5991-8133EN). If the two sets of instructions differ, use the instructions that are included in the box with the sealing gasket. Wiping of the whole processor can be performed as needed with compatible solvents/water. Question Q209 Does the rotameter help control flow for both Low and High Flow modes? Answer A209 No, the rotameter is only used for fine control for the Low Flow mode. When using Low Flow, the gas will run through low flow regulator and through the rotameter before entering the manifold to allow for more fine control at low pressure. When using High Flow, the gas will run directly from gas supply through the high flow regulator to the manifold. Question Q210 For the PPM-48, which switches correspond to which rows? Answer A210 As you are looking at the top of the manifold, the row furthest in the back (farthest from the front of the processor) will shut off the entire row A. The second switch from the back will shut off row B, and so forth. Therefore, the closest switch as you are looking at the front of the PPM-48 is for row D. The cartridge rack is labeled with Rows A-D. Row A is in the back, moving up to row D in the front. Question Q211 For the PPM-48, which switches correspond to which rows? Answer A211 https://www.agilent.com/en/support/switches Question Q212 For the PPM-48, why does my 1 mL cartridge rack have a black bar in between rows B and C when the other cartridge racks don’t have this bar? It is blocking my view of some of the cartridges. The bar has screws to hold it in place, so can I remove it? Answer A212 NO, you cannot remove this black bar. The bar is intended to provide additional stability to the 1 mL cartridge rack during compression. It is necessary to ensure proper pressure distribution between cartridges 1-12. The black bar sits between rows B and C making it more difficult to view cartridges that are in rows A and B. However, the bottom tips of the cartridges are still visible, and can be used to monitor flow of these rows if desired. Question Q213 How can I disconnect the processor from the nitrogen source? Answer A213 To remove the tubing from the back of the manifold (“gas inlet”), compress the outer ring of the push to connect fitting while simultaneously pulling out the tubing. Question Q214 How do I set the pressure for the processor? Answer A214 The processor has locking regulator knobs, one for Low Flow and another for High Flow. Pull out knob. Turn clockwise to increase pressure. Turn counter-clockwise to decrease pressure. When desired pressure is achieved, push knob in to "lock" pressure. It is easier to monitor/set the pressure when the flow rate selector is set to Low Flow or High Flow and the gas source is turned on. Question Q215 How do I set the pressure for the processor? Answer A215 https://www.agilent.com/en/support/set-pressure Question Q216 How do I use Captiva Syringe Filters? Answer A216 ??https://www.agilent.com/en/products/sample-preparation/sample-preparation-methods/filtration/stepbystep Question Q217 How frequently should I replace the manifold gasket on the processor? Answer A217 We recommended replacing them at least once a year for PM purposes. Under heavy use you might need to replace them once to 3 times per year depending on usage. We recommend daily inspections of the column seal for any signs of wear and recommend replacing if there are any visual signs of rip or tears as well as the column seal peeling off of the manifold. Question Q218 How frequently should I replace the manifold gasket on the processor? Answer A218 https://www.agilent.com/en/support/gasket Question Q219 How often should I replace my BHT-2 or BHT-4 gas filter? Answer A219 BHT-2 and BHT-4 traps are designed to filter 13 standard cylinders of gas. If you are unsure of the gas input quantity, a replacement interval of 1 year is recommended. Depending on usage, the filter can last up to 2 years. Question Q220 I can’t remove the back tubing on the positive pressure regulator, even when I follow the appropriate instructions. Answer A220 Sometimes the outer ring of the push to connect fittings can be dislodged. This may (but not always) be obvious visually, and will look like the outer ring is sticking out way more than the outer ring on a different push to connect fitting. One of the main causes of this dislodgment is trying to remove the tubing without pushing the outer ring in. Try re-seating the outer ring into the push to connect fitting with a pair of plyers by simply pushing the outer ring into the fitting (the tubing should be still connected- do not pull on the tubing while you are doing this). Once the push to connect fitting has been reseated, you should be able to push on the outer ring and simultaneously pull out the tubing. Question Q221 I see a pressure drop for High Flow pressure when I turn the flow selector from Off to High Flow. Is this normal? Answer A221 Yes, it can be. When the processor is in Off mode, there is no gas flowing through the system which can cause a slight back pressure to occur. When the flow selector is set to High Flow the gauge will immediately display the pressure that is currently being applied to the processor. For the PPM-48, with each row that is turned on via the row switches, there may be a slight decrease in the observed pressure in the high flow gauge. For example, if the high flow pressure is set to 80 psi with no rows on, the pressure will read 80 psi. As each row is turned on you may experience a 1-3 psi drop in pressure (if you are looking at the gauge). This is normal. However, if you are experiencing a much higher drop in pressure as row switches are turned on, it is most likely due to your input gas supply not meeting the minimum required pressure for the processor. Question Q222 I see a weird flow pattern when I am using 1 mL cartridges with my PPM-48. The end cartridges (columns 1-2 and 11-12) flow faster than the middle cartridges (3-10). Answer A222 Ensure that the 1 mL cartridge rack has the black stability bar installed and that it has not been removed. Without the stability bar, the cartridge rack may not compress correctly which can lead to inconsistent or variable pressure differences across the row. Question Q223 I’m struggling to remove the white paper backing of a new replacement sealing gasket. Answer A223 If the white paper backing of the replacement sealing gasket is difficult to remove- STOP immediately. One reason for the trouble may be because the adhesive is actually sticking to the paper backing and is being removed from the seal as you pull the backing off. The white backing when removed should be very smooth to the touch, and not sticky at all. You can use tweezers to help start to peel the backing off. Try a corner first, but if you are still having trouble, sometimes starting from the middle of the sealing gasket may work better. Question Q224 My sealing gasket won’t stick to the manifold. Answer A224 If the sealing gasket is old and currently installed: Replace the seal. If the replacement sealing gasket is not sticking initially: Ensure that all old adhesive has been removed from the manifold prior to attempting to attach a new sealing gasket. Old adhesive can be removed with acetone and/or isopropanol. A lint-free cloth should be used to prevent contamination to any of the manifold air holes. Avoid getting old adhesive into the manifold holes. If you notice adhesive in the holes, use small tweezers or similar item to remove all adhesive. Ensure that the white backing has been removed from the sealing gasket. The backing should remove clean (i.e. no adhesive residue should be on the backing). The back of the sealing gasket should feel sticky. Check to ensure that the adhesive remains on the sealing gasket as sometimes, the adhesive can be removed with the backing, rendering the sealing gasket not sticky! Press firmly to ensure the sealing gasket is attached. Follow specific instructions provided with the sealing gasket or in the manual to ensure proper installation. Question Q225 What are the recommended pressures for Low Flow and High Flow? Answer A225 Low Flow starting pressure recommendation is 3-4 psi, with the rotameter almost, but not fully, closed. If using High Flow for drying steps, any pressure between 50-80 psi is acceptable. Pressures can be set lower for High Flow for eluting more viscous samples. There is no pressure recommendation for this scenario as it will be sample dependent. Question Q226 What are the recommended pressures for Low Flow and High Flow? Answer A226 https://www.agilent.com/en/support/recommended-pressures Question Q227 What happens if I accidently lower or raise the manifold with the gas flow on (set to Low or High flow)? Answer A227 It is recommended to set the flow rate selector to Off when you are going to raise or lower the manifold. Movement of the manifold requires gas (it is all pneumatic). If you have the flow rate selector turned to Low Flow or High Flow then gas will be flowing through the manifold already. This is gas that is now not available to raise or lower the manifold. Therefore, to ensure enough gas is available to lower or raise the manifold, the flow rate selector should be set to Off during every compression/decompression step. Question Q228 What if I am using a harsh solvent in my sample preparation procedure? Answer A228 First, ensure that you have your processor installed in a location with proper ventilation, such as a fume hood, as dictated by the hazards of the solvent(s) being used. Also, ensure that appropriate personal protective equipment (PPE) such as gloves, googles, lab coat etc. are used according to the hazards. Regarding the effect of the harsh solvent on the processor, ensure that if any of the solvent spills on the processor or any of its components/accessories, it is wiped up immediately. Dispose of the waste and clean-up materials as dictated by the hazard and/or your laboratory protocols. It is especially important to wipe up any solvent that is on the cartridge rack, cartridges, or 96-well plate prior to compression. Prolonged exposure to solvent can damage the sealing gasket, so all efforts should be made to prevent direct contact with any solvents. Unnecessary, prolonged contact of the manifold (sealing gasket) and cartridge rack or 96-well plate should be avoided. Compression should occur only as dictated by the sample preparation method. Question Q229 What if I don’t have access to compressed nitrogen? Answer A229 The processor can operate using high purity filtered instrument air. Question Q230 What if I don’t have access to compressed nitrogen? Answer A230 https://www.agilent.com/en/support/n2 Question Q231 What is the charge capacity for SampliQ SCX 5982-3267? Answer A231 Our Polymeric SCX averages ~1,000 umol/g based on both elemental analysis and breakthrough method Question Q232 What is the rotameter and how do I use it? Answer A232 The rotameter is a device that measures the flow rate of a fluid or gas (in the case of the PPM it is measuring the GAS flow rate). The units for the rotameter are SCFH- standard cubic feet per hour- and will measure from 0-2.5 SCFH. The rotameter is only used with the Low Flow setting. Turning the rotameter knob counter clockwise will open the valve. Turning the knob clockwise will close/shut off the rotameter.DO NOT SHUT OFF THE ROTAMETER. Anytime the rotameter needs to be turned down/closed more, it is important to turn clockwise while ensuring you never fully close the valve. As soon as you feel any resistance, stop. It is a good idea to perform a slight counter turn to ensure it is not closed. Permanent damage can occur to the rotameter if the knob is closed regardless of whether the gas flow is on. Again, the rotameter should never be shut off regardless of gas flow being on or off. It is generally good practice to monitor the flows for the fastest flowing cartridges/wells and adjust pressure accordingly to ensure they are at the right level. As those cartridges/wells finish eluting, the rotameter can be opened more to allow more pressure to reach the slower flowing cartridges/wells. Question Q233 What other parts or accessories do I need to order with my PPM-48? Answer A233 Regardless of sample preparation technique, a cartridge rack(s) will need to be purchased (3 options). Cartridge racks: 1 mL cartridge rack p/n 5191-4102 3 mL cartridge rack p/n 5191-4103 6 mL cartridge rack p/n 5191-4104 A collection rack(s) (to hold test tubes or autosampler vials) will also be necessary. Collection racks: 10 × 75 mm tubes p/n 5191-4105 12 × 75 mm tubes p/n 5191-4106 13 × 100 mm tubes p/n 5191-4107 16 × 100 mm tubes p/n 5191-4108 12 × 32 mm autosampler vials p/n 5191-4109 If you are doing SPE, then you most likely have all or some of the following steps PRIOR to elution in which you will not want to collect the eluting liquid: conditioning, equilibration, loading, washing. Unless you wish to collect the liquid from these steps, then a waste rack with waste bin will be very useful. The waste rack and a 3/pk of bins are included with the initial purchase of a PPM-48; however, additional waste bins (3/pk: p/n 5191-4113) may be purchased. It is recommended that you use a gas filter to filter your gas supply. A BHT-4 (1/4 fittings) or BHT-2 (1/8 inch fittings) are available from Agilent. Sealing gasket is a consumable that will need to be replaced (1-2x per year), however Agilent doesn't recommend ordering this at the same time as the processor order due to the expiration of the sealing gasket adhesive. Question Q234 What other parts or accessories do I need to order with my PPM-96? Answer A234 Most of the accessories/parts that you will need with the PPM-96 will just be the sample preparation 96-well plate and 96-well collection plate you require for your method. The single well waste plate that is provided with the initial processor order can be reused, but additional can also be purchased as necessary (p/n 5191-4121). If you are planning on using the PPM-96 for tables 1 mL cartridges, a 1mL tables cartridge holder is required (p/n 5191-4119). The sealing gasket is a consumable that will need to be replaced (1-2x per year), however Agilent doesn't recommend ordering this at the same time as the processor order due to the expiration of the sealing gasket adhesive. It is also recommended that you only order as needed, and do not keep lot of spares on hand due to the expiration of the sealing gasket adhesive. Question Q235 What size frits are used in the SampliQ products? Answer A235 The pore size of the frits is 20um. More SampliQ-specific FAQs and answers can be found in the attached document. Question Q236 When I first turn on the flow (Off to Low Flow) I get a large burst of pressure which makes my samples elute at an undesirable rate for the first few seconds. How can I prevent this? Answer A236 This "burst" is usually due to backpressure build up from a period of processor inactivity. A slight build-up is normal. To avoid this, you can "prime" the processor. If using Low Flow mode, simply turn the flow selector from Off to Low Flow prior to compression of the manifold for 2-3 seconds or until pressure stabilizes and returns to original setting. Ensure that the flow selector is returned to Off prior to compression. If using High Flow mode, the same procedure can be followed with the flow selector set to High Flow. However, usually this difference in High Flow is nominal since the pressure is already so high. AKA- the difference between 3 and 5 psi (Low Flow) can be substantial for your samples, when the difference between 80 and 83 psi (High Flow) generally won't significantly affect the flow rates of the cartridges. Question Q237 When I turn on my gas supply source, the processor pressure gauges don’t register any pressure. Answer A237 Disconnect the gas supply tubing from the processor. Turn on the gas supply source and ensure there is flow coming from the installed gas supply tubing. Turn off the gas supply. Insert the tubing by pushing it into right-hand port ("Gas Inlet") of the processor (when looking at the back of the unit). When the tubing is installed properly, you should not be able to pull it out of the port. Then turn on the nitrogen source. Question Q238 Where can I find the QuEChERS SOP (Standard Operating Procedures)? Answer A238 We have a recommended SOP that can be followed using the AOAC 2007.01 and the EN Method 15662 kits. Question Q239 Why can’t I compress/lower the manifold? Answer A239 Ensure the nitrogen source is turned on, and the flow rate selector is set to Off. Ensure the processor is receiving adequate gas (60-100 psi) by checking the pressure gauges on the unit. If you are still unsure of the pressure or if the processor is receiving gas, use a pressure gauge/regulator with your gas supply to monitor. Both compression switches must be pressed and held simultaneously to lower or raise the manifold. If you let go of the switches too early, the manifold will raise back up. This in an intentional safety feature. The manifold will stop when it reaches the cartridge or plate stack. Make sure you hold down the switches until this contact has been made. It doesn't hurt the processor to hold the switches for an extra second or two after compression to ensure the manifold is compressed. Question Q240 With what gas supply pressure(s) is the processor compatible? Answer A240 The processor can operate with input pressures between 60 psi (minimum) and 100 psi (maximum). Optimum source pressure is 80 psi. If you are unsure of the pressure of your in-house gas, it is highly recommended that a supply gas pressure gauge and/or regulator is installed.
Syringes and Sample Introduction
Question Q241 Do econofilter syringe filters come with a certificate? Answer A241 No. Certificates are not available for econoline vials. Question Q242 How do I choose the right syringe filter for my application? Answer A242 https://filtrationselectiontool.chem.agilent.com/product-guide/agilent-filters-product-guide/ Question Q243 How do I find needle gauge dimensions of a manual syringe? Answer A243 https://www.agilent.com/en-us/products/lab-supplies/chromatography-spectroscopy/more-information Question Q244 Syringe selection guide Answer A244 https://www.agilent.com/search/gn/syringe-selector?searchTermRedirect=Syringe%20selection%20guide Question Q245 What are the steps to change the sample needle? Answer A245 https://www.agilent.com/en-us/support/gas-chromatography/kb002362 Question Q246 What does the diameter of the filter refer to? Answer A246 The diameter of the filter refers to the active membrane diameter, not the filter body. Question Q247 What ships with the Color Coded Manual Syringes?: Answer A247 The Manual Syringes ship with an insert that describe how to use the packaging as a syringe holder and general care and use instructions. It also provides instructions for getting the Certificate of Conformance. Question Q248 What syringe offerings does Agilent have? Answer A248 https://www.agilent.com/en-us/products/lab-supplies/chromatography-spectroscopy/syringes#1 Question Q249 What types of gas chromatography syringes are available? Answer A249 https://www.agilent.com/en-us/support/gas-chromatography/kb001569 Question Q250 Will the syringe filters work with any manufacturer's syringe? Answer A250 The syringe filter has a standard luer connection that should fit any syringe using a standard luer connection.
Tools, Fittings, Capillaries, Tubing, and Supplies
Question Q251 Can an oil free Rough pump be used With a Diffusion pump MS? Answer A251 https://www.agilent.com/en-us/support/mass-spectrometry/kb004804 Question Q252 How are the Markes Difflock caps (MKI-MTD-1204) packaged and what color are they? Answer A252 The caps are supplied in packages of 10. They have an inert coating which gives them a unique color/appearance. Question Q253 The UI gold seal has spots – is it used, or rusty? Answer A253 https://www.agilent.com/en-us/support/gas-chromatography/kb005987 Question Q254 UltiMetal Plus Flexible Metal Ferrule Answer A254 https://www.agilent.com/en-us/support/gas-chromatography/kb005973 Question Q255 What is the Gold-seal with the cross used for? Answer A255 https://www.agilent.com/cs/library/usermanuals/public/Tips%20Tricks%20of%20Injector%20Maintenance_012810.zip Question Q256 What is the interval for Rough pump oil replacement? Answer A256 https://www.agilent.com/en-us/support/mass-spectrometry/kb002041 Question Q257 What tool should I use for cutting fused silica columns? Answer A257 https://www.agilent.com/en-us/products/gas-chromatography/gc-columns/capillary-tubing-accessories/fused-silica-tubing-cutters Question Q258 What traps are available for the Purge and Trap device? Answer A258 Information can be found in the consumables catalog under Teledyne Tekmar Purge and Trap supplies https://www.agilent.com/cs/library/catalogs/public/5991-5213EN_GC_Catalog_Supply.pdf Question Q259 What tubes and accessories are available for the Markes Thermal Desorption system? Answer A259 Information can be found in the consumables catalog under Markes Thermal Desorptionhttp://www.agilent.com/cs/library/catalogs/public/5991-5213EN_GC_Catalog_Supply.pdf Question Q260 Where can I find information about Ultimate Unions? Answer A260 https://www.agilent.com/en-us/products/gas-chromatography/gc-supplies/ferrules-connectors/ultimate-union
Vials and Closures
Question Q261 Can I use septa on the HPLC vials and test tubes when using… Answer A261 https://www.agilent.com/en-us/support/dissolution/kb005930 Question Q262 How much can a sample vial be filled? Answer A262 https://www.agilent.com/en-us/support/gas-chromatography/kb005143 Question Q263 What are 'Certified’ consumables? Answer A263 Agilent Certified consumables come with a certificate of analysis, so you can be sure that they will perform even in the most demanding environments.