Good afternoon everyone,

I am using the following septum for SPME analysis:

The analysis of a blank vial (empty) will give the following chromatogram in GC-MS:

The first peak is octamethylcyclotetrasiloxane (Si4) and the others are Si5, Si6 and Si7.

Because octamethylcyclotetrasiloxane is a 'Reproductive toxicity 2' compound, it would be very interesting not to have it leaching out of the septum. Especially when it is a compound of interest for me.

Do you know another type of septum more inert?

Thank you.

Parents
  • Normally the cyclic siloxanes only come off at very high temperatures or by prolonged contact with solvents or solvent vapours, especially DMSO, DMF, THF, and others more aggressive solvents; not with a single injection from a headspace vial under classical conditions. And they very often do not come from your vial septum, but the GC inlet septum.

    You can try to use a PTFE/butyl rubber disk/septum like 8010-0140 for the vials and if you still see the peaks, it is from the inlet septum.

  • I planned to try the PTFE/Butyl septa (8010-0140 | Agilent) received but it did not go as expected.

    The metal housing of the fiber got bent, when using a CTC, destroying the SPME fiber assembly.

    The septum is way too hard to be pierced by the blunt tip of the SPME. The fiber installed was new.

    I even tried by hand with a cone-tip 10 mL syringe and I could feel the difference when compared to a PTFE/Silicone septum.

    Because SPME fibers are not cheap, I will not try again.

    I did not find anything on the page of Agilent about tip compatibility with this septum.

    We are using SPME fibers from Supelco (ref.: 57319).

Reply
  • I planned to try the PTFE/Butyl septa (8010-0140 | Agilent) received but it did not go as expected.

    The metal housing of the fiber got bent, when using a CTC, destroying the SPME fiber assembly.

    The septum is way too hard to be pierced by the blunt tip of the SPME. The fiber installed was new.

    I even tried by hand with a cone-tip 10 mL syringe and I could feel the difference when compared to a PTFE/Silicone septum.

    Because SPME fibers are not cheap, I will not try again.

    I did not find anything on the page of Agilent about tip compatibility with this septum.

    We are using SPME fibers from Supelco (ref.: 57319).