Run a partly used ScreenTape on another TapeStation

I have a ScreenTape of which 3 lanes have been used on one of the TapeStations we have in the building. Now I wish to use this same ScreenTape on a different TapeStation, but this ScreenTape is seen as ‘unused’ by the other TapeStation.

Is there a way that I can still use the partly used ScreenTape on a different TapeStation than it has been running? (Some software settings or letting the used lanes run with sample buffer + water?)

  • If you are in the US please contact Electrophoresis Support directly at 1800-227-9770 > 3 > 4 > 1 or electrophoresis@agilent.com.   

    Otherwise, please reach out to your local customer care: https://www.agilent.com/en/contact-us/page

  • Hello! We would like to share with you the news that we now have a dedicated area for our Automated Electrophoresis products.  Please come visit our space in Agilent Community!  Thank you, Shweta

  • Lane usage is stored in the PC. Merging the databases between the 2 systems is not possible at the moment. If you want to try to use the partially used ScreenTape, please consider skipping lanes already used on the ScreenTape. Please be careful not to use previously used lanes as this will lead to inaccurate results.

    Since your ScreenTape has 3 lanes that have been previously used, please leave the first 3 wells blank.

    Example: ScreenTape has 13 lanes remaining because 3 were previously consumed (recorded on a different PC). Now you want to run 6 samples using a different PC. Highlight 9 wells to run in the Controller software. Do not add sample to the first 3 wells of the strip tube (you may leave these blank).

    Please note that this is possible only when running samples from the strip tube location. This could be applied when running samples from the plate location only if an electronic ladder is used because when a run ladder is used the ladder is run from a fixed location (strip tube location A1) and always loaded in the first available lane. 