Tapestation 4150 Needle Replacement Jamming

I'm replacing the needle on my Agilent Tapestation 4150 and every time I run it, it jams.

I'm following the procedure and I'm not sure what I'm doing wrong

Parents
  • Thank you for contacting the Agilent support team. Although I am sure you have the correct information to perform needle exchange, reiterating below to be sure.

    Part # 5067-5783 is needed for 4150 TapeStation needle exchange.

    It is recommended to first reboot the entire system in the suggested sequence before initiating the exchange. Specifically, turn off both the instrument and PC. First, turn the PC ON but do not open controller software (SW). Then turn ON the instrument. Once the instrument is initialized, launch the controller SW, we expect connection to be established between the instrument and PC.

    Needle cartridge can be exchanged from the controller SW > Tools > Maintenance > Change needle following detailed steps starting page 124 of the 4150 TapeStation manual. It is critical to wait approximately two minutes and not to turn off the instrument at this step to allow needle calibration to be completed.

    In reference to the jam/issue you are facing, please email us the following information at electrophoresis@agilent.com :

    1. When does the block error come, for example, right after you click, proceed etc?
    2. Screenshot of the error message.
    3. Instrument log files. To collect those, please close the controller SW. Launch Analysis SW>File> Help> Export Log Files> email us the .zip folder created.

    Thank you, Shweta

Reply
  • Thank you for contacting the Agilent support team. Although I am sure you have the correct information to perform needle exchange, reiterating below to be sure.

    Part # 5067-5783 is needed for 4150 TapeStation needle exchange.

    It is recommended to first reboot the entire system in the suggested sequence before initiating the exchange. Specifically, turn off both the instrument and PC. First, turn the PC ON but do not open controller software (SW). Then turn ON the instrument. Once the instrument is initialized, launch the controller SW, we expect connection to be established between the instrument and PC.

    Needle cartridge can be exchanged from the controller SW > Tools > Maintenance > Change needle following detailed steps starting page 124 of the 4150 TapeStation manual. It is critical to wait approximately two minutes and not to turn off the instrument at this step to allow needle calibration to be completed.

    In reference to the jam/issue you are facing, please email us the following information at electrophoresis@agilent.com :

    1. When does the block error come, for example, right after you click, proceed etc?
    2. Screenshot of the error message.
    3. Instrument log files. To collect those, please close the controller SW. Launch Analysis SW>File> Help> Export Log Files> email us the .zip folder created.

    Thank you, Shweta