ICP-OES sample digestion

Hi All,

I was wondering if anybody knows of any good go-to resources for sample digestion methods? Any ideas are appreciated. I have currently looked through ASTM digestion methods but couldn't find anything suitable for our application.

On that note, we have an aramide polymer we want to analyse for S content, but I am concerned about the loss of analyte with H2S evolving after microwave digestion. If anyone with experience with sulfur analysis could chime in on whether or not this is a genuine issue I would be grateful :-)

Cheers!

Parents
  • Hi,

    We use a CEM SPD instrument for acid digestion. When I want to digest an unusual (for me) sample I always look first in the CEM.com website. Under resources/ method notes there are a lot of different products listed with acids used plus temperatures. Maybe you can tweak something from there for your specific instrument.

    I analyse a lot of sulphur in oleochemical samples, and I always digest 500mg in 10 mL nitric acid in a closed vessel SPD system. I never loose any sulphur. However, I think this depends greatly on the initial compound in your sample. In the past I once did a few different spikes of sulfuric acid together with my sample in a microwave digestion, and the recovery was 0% Grin.

    I analyse sulphur on mass 32 or 34 with xenon as a collision gas. Because I am forced to analyse 100x diluted samples (because of sample prep) my quantification limit is something like 2/ 3 ppm S. For digested oleochemical I use an internal standard solution with added IPA. This will enhance the S sensitivity.

    Hope this will help you a little.

    BTW I use an ICP-MS!!

  • Thank you! Seems like a very useful website. Glad to hear there aren't normally issues with sample loss - your acid volumes are a bit higher than what we have been using but I think we can afford the extra dilution in exchange for the greater volume to dissolve the sulfur in. Will try a spike test and see what the results are.

Reply Children
No Data
Was this helpful?