Hello all,
How could I do analysis of Silicium in SIO2 in pharmaceutical ingredient an ICP-MS? What is the tips to prepare the sample?
There is a spcecial care in equipment?
Do you have any application note from Agilent?
Hello all,
How could I do analysis of Silicium in SIO2 in pharmaceutical ingredient an ICP-MS? What is the tips to prepare the sample?
There is a spcecial care in equipment?
Do you have any application note from Agilent?
Hello,
Si analysis using ICPMS requires better H2 mode, as Si has known interference from N2 and CO. So this your best choice for Detection Limits. I do not see any challenge in preparing your samples and you can use your standard quartz sample introduction system.
Hello,
Si analysis using ICPMS requires better H2 mode, as Si has known interference from N2 and CO. So this your best choice for Detection Limits. I do not see any challenge in preparing your samples and you can use your standard quartz sample introduction system.