How to get the absorbance 0.9 when measuring a solution of Cu 5mg/L with a flame AAS system.

Hello,

We recently set up an atomic absorption spectrophotometry with a flame system. We need to test the performance of this new system by measuring the absorbance of a copper solution 5 mg/L.

However, we only obtained the maximum absorbance is 0.6. On the other hand, we found on the website of Agilent, the absorbance of Cu solution 5 mg/L is typically greater than 0.9 Abs running with a Mark 7 Atomization system.

Has anyone obtained the absorbance that is greater than 0.9 when measuring a solution of 5 mg/L Cu, and could you please share with us optimal conditions to get such a high absorbance.

Thank you very much in advance.

Parents
  • Hello, with the  AA instruments. It can: achieve high sensitivity—typically 0.9 Abs. from 5 mg/L Cu; achieving these values needs to be considered; you need to consider the quality of acetylene, air free from moisture and water contents, additional optimisation of sample uptake volume, and posting of the impact bead; BTW, did you remove the mixing paddle from the spray chamber?

  • Hi DDU,

    We brought acetylene from a trustable provider.
    We also adjusted the sample uptake volume by turning the nebulizer.
    Today, we tried to the Cu solution of 5 mg/L both when having the mixing paddle and removing the mixing paddle.
    The maximum absorbance was only 0.6 with the mixing paddle. When removing the mixing paddle, the absorbance could only be more than 0.9 if we changed a sample capillary with a larger diameter.
    When we used a standard sample capillary provided by the manufacturer, the absorbance was only 0.6.
    Because there is a sentence on the website of Agilent indicating that “the absorbance of Cu solution of 5 mg/L is typically greater than 0.9 Abs running with a Mark 7 Atomization system”.
    So, could I ask that can this value be obtained in normal or specific conditions?
    Because this value is reported on the website, I think it would be better to include conditions at which the value has been obtained.

Reply
  • Hi DDU,

    We brought acetylene from a trustable provider.
    We also adjusted the sample uptake volume by turning the nebulizer.
    Today, we tried to the Cu solution of 5 mg/L both when having the mixing paddle and removing the mixing paddle.
    The maximum absorbance was only 0.6 with the mixing paddle. When removing the mixing paddle, the absorbance could only be more than 0.9 if we changed a sample capillary with a larger diameter.
    When we used a standard sample capillary provided by the manufacturer, the absorbance was only 0.6.
    Because there is a sentence on the website of Agilent indicating that “the absorbance of Cu solution of 5 mg/L is typically greater than 0.9 Abs running with a Mark 7 Atomization system”.
    So, could I ask that can this value be obtained in normal or specific conditions?
    Because this value is reported on the website, I think it would be better to include conditions at which the value has been obtained.