I am running ICP-MS 7700 with mass hunter, for tune with Helium gas mode ON, my %RSD is little bit high. can anybody help me to make it pass? what will be the troubleshoots? I cleaned cones, changed tubings for perilistic pumps still no improvement.
I am running ICP-MS 7700 with mass hunter, for tune with Helium gas mode ON, my %RSD is little bit high. can anybody help me to make it pass? what will be the troubleshoots? I cleaned cones, changed tubings for perilistic pumps still no improvement.
Hi SSaiyad,
1. In helium mode. If He mass count improved with Zero helium flow. then please check following.
2. If above mentioned points are good i suggest octupole replacement.
Hi SSaiyad,
1. In helium mode. If He mass count improved with Zero helium flow. then please check following.
2. If above mentioned points are good i suggest octupole replacement.