Hi All,
I am just wondering ,why i can't select the intelligent rinse function when using an AVS 7?
Am i overlooking somehting or is this functionality just not available with an AVS?
Hi All,
I am just wondering ,why i can't select the intelligent rinse function when using an AVS 7?
Am i overlooking somehting or is this functionality just not available with an AVS?
The AVS is constantly rinsing when not integrating the sample signal. Typically, you shouldn't need the Intelligent rinse feature because of this. If you have a sticky element, you can increase the rinse time in the AVS parameters on the Conditions tab.
If you're still getting carry-over with the AVS, you may need to consider an acidic rinse solution, or increasing the % of acid in the rinse solution.
thank you for the quick reply! I understand that the AVS is constantly rinsing, it would just be helpful if the rinse duration was adjusted on a per sample basis.
thank you for the quick reply! I understand that the AVS is constantly rinsing, it would just be helpful if the rinse duration was adjusted on a per sample basis.